I am sorry if this has been discussed previous... I searched and didn't find anything.  Quick question...

We are transfusion service that performs DAT's using poly-clonal IgG... if it is positive, we run the mono-clonal IgG, however, we do not run the C3d.    How many of you would and/or do run the complement control cells for DAT QC in addition to Check Cells?  

 

Share this post


Link to post
Share on other sites


With a positive poly on a DAT, we only run the anti-IgG here.  But we send the specimen out to a nearby lab for the Compliment.  We do not do the anti-compliment as we do not want to pay for the QC material that we would have to buy (and mostly waste when it outdates).

Scott

 

Share this post


Link to post
Share on other sites

TRM.40200 DAT Controls Phase II
When performing an antiglobulin test with anti-IgG or polyspecific antiglobulin reagents, IgG-coated red blood cells are used as a control in all negative antiglobulin tests. NOTE: IgG-coated red blood cells must be used to confirm all negative antiglobulin test results when the antiglobulin reagent used for testing has anti-IgG reactivity. Tests found negative by tube methodology must be verified by obtaining a positive test result after adding IgGcoated (control) red blood cells. If a licensed blood typing system is used that does not require
verification of negative test results using IgG-coated red blood cells, an appropriate quality control procedure must be followed, as recommended by the manufacturer.
Evidence of Compliance:
✓ Records of testing that include control results confirming negative antiglobulin tests

TRM.40210 DAT Phase II
When performing an antiglobulin test with anti-C3 antiglobulin reagents, C3-coated red
blood cells are used as a control in all negative antiglobulin tests.
NOTE: Complement-coated red blood cells must be used to confirm all negative antiglobulin
test results when the antiglobulin reagent used for testing has anti-C3 reactivity. Tests found
negative by tube methodology must be verified by obtaining a positive test result after adding C3-
coated (control) red blood cells.
If a licensed blood typing system is used that does not require
verification of negative test results using C3-coated red blood cells, an appropriate quality control
procedure must be followed, as recommended by the manufacturer. If a polyspecific antiglobulin
reagent is used, refer to checklist item TRM.40200.
Evidence of Compliance:
✓ Records of testing that include control results confirming negative antiglobulin tests

**********************************************************************************************************************************************

I was cited for this years ago. I called CAP and was told that because poly AHG has anti-C3 reactivity as well as anti-IgG reactivity, both had to be confirmed. In addition to this std, she also referred me to the all common checklist which requires that we perform QC on reagents every day of use. So unless the manufacturer of our reagent had some other recommended QC procedure for C3 reactivity, we were required to use the complement coated cells. I put a standing order in for C3 coated cells that day, sent the confirmation email to CAP and my citation was considered corrected on site. I would assume that AABB would view this in a similar way, not to mention CLIA.

When we do a DAT, we are looking for both anti-IgG and anti-C3 activity.  If the DAT is positive with poly and anti-IgG, that doesn't preclude anti-C3 activity. If you aren't doing QC for the anti-C3 activity of your poly AHG, how can you demonstrate that your reagent is reacting properly? If you send all DATs out to check for C3 activity, then you would only have to QC the anti-IgG activity and your reference lab would be responsible for the C3 activity. 

Having said all that.....have I ever seen a failure with the C3 activity? Nope and I don't expect to. I've given students anti-C3b, -C3d reagent that's outdated by years and it still works just fine. But that's irrelevant and not how the game is played. We don't do very many DATs, but that's also irrelevant. So, I stock the C3 coated cells. Cost of doing business. I find ways to save in other areas.

Share this post


Link to post
Share on other sites

I have never seen that interp for polyahg requiring IgG and C3 sensitized cells.  It seems to me that the final sentence of referring to TRM.40200 allows the use of the IgG sensitized cells to document the reactivity of your poly reagent.  I know I am not the only one who interprets this.  I will f/u w CAP on this.

Share this post


Link to post
Share on other sites

Regardless of a particular regulatory requirement concerning DAT QC,  I believe that one must at the least follow manufacturer's recommendations for a particular system.  For our Ortho BioClone polyspecific, the procedure requires both types of sensitized cells for positive controls.  This makes sense to me.  And besides showing that the cell-washer is working properly, one has to prove that the reagent can give a positive with both compliment- and IgG-sensitized cells.

Scott

Share this post


Link to post
Share on other sites

I have heard from CAP on this question - and I quote:

" Mr. Saikin

If you are using polyspecific reagent, you are only required to confirm negative results with IgG-coated red cells.  C3-coated red cells are only required with anti-C3 reagents are used (sic).  The last sentence in the note for TRM.40210 states that if polyspecific reagent is used, the TRM.40200 applies.  I hope this helps.

Regards,

Sarah Fabian, MLS(ASCP)"

If you use polyahg, you only need to run IgG check cells to confirm negatives.

Share this post


Link to post
Share on other sites

But...but...but...

polyspecific reagent DOES contain anti-C3...

(And I do not think that Ms. Fabian would suggest ignoring any manufacturer's recommendations contrary to her quote above.)

 

Scott

Share this post


Link to post
Share on other sites

My interpretation: Users of POLYSPECIFIC antiglobulin reagents are obliged to verify performance each day of use, i.e., QC should involve use of IgG-coated cells AND Complement-coated cells. This gives the user confidence that the reagent is performing as expected.

During routine testing, addition of IgG-coated cells to negative tests is sufficient to verify that the IAT was performed correctly - correct/effective washing, the antiglobulin reagent was added and is reactive, etc.

If it were a requirement to add IgG-coated cells and complement-coated cells to every negative IAT using polyspecific antiglobulin, it would be necessary to run everything in duplicate - one set would get IgG-coated cells and the other set would get complement-coated cells. I don't think that is the case.

Edited by exlimey
Typo

Share this post


Link to post
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now

  • Similar Content

    • By Jermin
      Hi All,
      I have a question, but firstly good old story time for some context. I came across a patient who had positive antibody screen on all three screening cells used (BioRad). I was concerned this may be an auto and pan-reactive, and required units. Performed a monospecific DAT, showing a positive reaction to IgG only. By this time antibody panel finished cooking and showed the patient may have anti-Fya , but couldn't do phenotype. By this time I was nearing my shift so handed it over to my colleague and asked for some units to be crossmatched. However, he refused as DAT was positive and said he rather send the sample to reference laboratory for them to crossmatch. The next day I crossmatched units to verify if it could have been done in our laboratory (just because I am sad that way), and turn out the unit I crossmatched was compatible (which I wasn't surprised about)
      Question
      Why does positive DAT (or the cause of positive DAT) sometimes interfere with IAT techniques (such as antibody panel and crossmatch) and sometimes it does not? If both use AHG, then wouldn't positive DAT with IgG cause antibody panels shows pan-reactive with red cells? But obviously it doesn't, but I'm trying to figure out why, and I'm sure the answer is quite obvious. 
      My laboratory seems very hesitant whenever they see anything regarding autoantibodies or positive DAT, and thinks that sample cannot be crossmatched in-house and needs to be sent off without even trying to investigate. Hopefully, by me asking this question, I can explain it back to my colleagues (but obviously take all the credit).
      Cheers in advance,
      Jermin
       
    • By Jermin
      Hi All,
      Its been a while since I came back to this forum, but glad I feel like I have gained a lot more insight. I feel like I'm a bottomless cup. 
      So I come with a question, for which there is not going to be a definitive answer (but with BB, is there ever one?), but hopefully, I would gain a bit of understanding.
      Background: So, our laboratory has started sending samples to reference laboratory for genotyping of the foetus by FDNA, which is great, since we would figure out the Rh(D) status of the baby (on most occasions) before they are born! So our laboratory has set up a flow chart which basically mentions that you do not need to send cord sample of Rh(D) negative mother if the baby is shown to be Rh(D) positive (or D positive, I am quite wary when trying to talk about Rh group), and only send cord if baby of Rh(D) negative mother if the FDNA shows the baby is Rh(D) negative, just to confirm the accuracy of FDNA. It sounds kinda counterintuitive, but we will soon be just not processing any cord sample for the ones we performed FDNA on. That means no cord Blood Group or DAT on a lot of post-delivery patients.
      Question:  By missing out DAT, we would possibly be missing out on detecting ABO incompatible HDN. How significant do you think it is in the early stages? Is it OK to wait to see if the patient shows signs of jaundice and for them to send a DCT sample afterwards? 
      Bonus Question: What does your Hospital/Laboratory do in the event of positive DAT on cord sample, and why do you do it?
      I had a read through one of the articles stating about the significant of DAT, but they called the Rh blood group as Rhesus, so I'm not going to take them too seriously

      Cheers,
      Jermin
    • By MaryPDX
      Has anyone used this card? What did you think of it?  I know it's not available in the US yet, but I'm looking forward to it. 
       

    • By Darren
      It's been years since I've done test tube blood banking and even then I didn't have to do any sort of daily qc. I'm bringing back the option of running a test tube IS xm to speed up add on units and I don't know if there's any sort of QC required for this. It's patient vs donor. Is there anything to QC daily?
      Thanks,
      darren
    • By Darren
      I run the Provue with the polyspecific DAT card. We only run an IgG control on this. I think I want to transition to the card that does IgG and Complement separately. What control cells do everyone use for the anti-complement gel testing? Brands and such, I mean, I know how to run them.