    Everwhere I have worked uses expired panels to rule out/in antibodies.  The cost of keeping enough in-date panels to work up all anitbodies would be enormous.  AABB, CAP and FDA have all inspected the

  • Malcolm Needs
    Malcolm Needs

    We use known, very weak, antibodies of the same specificity as we are trying to rule out. For example, if we are trying to rule out a weak anti-K, which only reacts with presumed homozygous K expressi

comment_50038

Yes, but with a very strict control regimen.

  • Author
comment_50039

What kind of control regimen do you use? Our sister hospital just got cited from CAP for not performing QC on the expired panel of use and are not sure what to use for QC.

comment_50043

We use known, very weak, antibodies of the same specificity as we are trying to rule out. For example, if we are trying to rule out a weak anti-K, which only reacts with presumed homozygous K expression, we use a known, very weak anti-K to ensure that the K antigen on the expired ABID red cells has not degraded.

comment_50051

Malcolm - are these the weak abs you spoke about earlier? . . . provided by your nat'l service.

We do not have that luxury in the USA.

I use a 1:4 dilution of either Fya or Fyb: have not had to deal with Fya-b- cell yet.

comment_50055

Well, some are David (anti-D, anti-c, anti-K and anti-Fya), but others are just some that we have collected from patients over the years and frozen down for just such an eventuality.

comment_50057

We keep panel cells for 4 months. Our SOP states they are acceptable until they start to hemolyze or turn brown. We have never been cited by AABB for using them.

comment_50058

Blow me! 4 months seems a bit long to me, even in CellStab!

comment_50060

We keep our panels until they start to hemolyze. We do run QC on all our expired panels. We use Anti-Fya typing sera with a heterozygous positive Fya cell and a Fka negative cell. we have never been cited by CAP for this.

comment_50062
Blow me! 4 months seems a bit long to me, even in CellStab!

That made me burst out laughing. In America that would be a very vulgar statement indeed !

comment_50064

Here is an ongoing question regarding panel cell QC. If you QC your expired panels cells, do you QC your in date panel cells? For those of you that use Anti-Fy, what if the antibody you are trying to ID is not a Fy? To be truly accurate with QC, would you not need to test every cell for every antigen listed? Haven't we all found examples of antibodies that haven't read the book and do not react with all the cells they are supposed to? Is it because of the antibody or because of the cell? Would QC help in this situation?

Just playing devil's advocate :devilish:

comment_50068
That made me burst out laughing. In America that would be a very vulgar statement indeed !

Oooops. Sorry. What is it? Friends divided by a common language - or something along those lines!!!!!!!!!!!!!!!!!!

comment_50074
Does anyone use expired ABID Panels to help rule out cells homozygously?

We use commercial antisera as positive control and saline solution as negative control.

comment_50084

We do not use expired cells to rule out, only sometimes to prove the specificity (extra positive).

But when you do it you have to do the controls that Malcolm sugested, show that there is weakening of the antigen you use to rule out. In the absence of a weak antibody you can switch to a diluted antibody or perform the antigen expression in titer compaired to a cell that is not expired. (a weak antibody is better, but this can be used as an alternative).

The "4 months rule" did not make me say the same as Malcolm :D but I needed a few minutes to get from the flour on my chair again.

Peter

  • 3 weeks later...
comment_50360

Malcolm, I have tried to come up with a way to explain the phrase, but words that will not get me in trouble elude me. :redface:

comment_50361

Which phrase Elizabeth? - send me a private message if it is that embarrassing (or email me on Malcolm.needs@blueyonder.co.uk).

comment_50376
Which phrase Elizabeth? - send me a private message if it is that embarrassing (or email me on Malcolm.needs@blueyonder.co.uk).

With respect to 'Blow me'; this would be the same as 'That's a funny thing' or 'Blow me down', generally 'What a surprise!'

comment_50418
Oooops. Sorry. What is it? Friends divided by a common language - or something along those lines!!!!!!!!!!!!!!!!!!

Malcolm, in the U.S. it's a euphemism for oral sex.

---

We QC our expired panel cells with the corresponding commercial antisera to whatever antigen we are using that particular cell to rule out with. (Example: If using an expired K+ cell to rule out Anti-K, we test the cell with commerical Anti-K)

comment_50424

Om my goodness. I'm sorry if I have caused any offence. It doesn't mean that over here.

comment_50431

I'm sure no one is offended. We are all developing a greater global awareness, right? :)

Last I knew no one in the UK would use the nickname Randy for someone named Randall, but that is common in the US.

Edited by Mabel Adams

comment_50528

Another QC Method I have used is: On the outdated panel, run a cell that is positive (and reactive) for the "known" antibody and look for a similar strength of reaction. For example, let's say the patient has Anti-Jka reacting 2+ (indated panel), Anti-C reacting 1+ (indated panel), and you are using your outdated panel to rule-out Anti-S

(so would need to run a cell that is Jka-, C-, S+)....you would also test a cell that is Jka+ or C+ and see if the strength of reactivity on the outdated panel is comparable to the strength of reactivity on the indated panel. That would be your control cell.

Just a thought...

Brenda Hutson, CLS(ASCP)SBB

comment_50535

Just because the Jka or C antigens were up to snuff, does that mean the S antigen would still be detectable?

comment_50544

We don't know that for sure on our in-date panels either. Someone may have spilled bleach nearby and destroyed the S antigen in some vials. The only way to be totally sure would be to QC every antigen on every cell every day of use.