Gel and tube discrepancy in antibody screen


Recommended Posts

hello, guys i'm having an issue with an antibody screening results of the donors 

i had an issue with couple of donors the first one he's an O+ , on the gel it gave negative ICT, but in the tube method it gave a positive P3 , identification is negative on tube method 

second donor was positive on the gel method, he's an A- , and the tube ICT was negative , tube identification was negative 

bio rad reagent used in the tube method 

and ortho diagnostics in the gel 

what's the possible causes of these results 

thanks 

Edited by radwan1411
Link to comment
Share on other sites

In first scenario, Gel negative but tube ICT positive. Was the ICT positive on all cells or just some cells? Did you repeat both Gel and ICT screens to make sure there were no test performance errors. Did you test with with a different set of reagents to make sure reagents had not been contaminated?

In second scenario, basically same questions. However if you have reactions in gel with all cells then it might be an Ortho Enhancement Isoagglutination caused by the preservatives in the reagent cells. You can try repeating the antibody screen in gel using reagent cells that have been washed first to remove the additive.

Link to comment
Share on other sites

I find that Gel screens are positive and tube are negative when there is either a Cold Antibody or Rouleaux.  Sometimes the gel finds an HLA that may not be picked up in tube either.  In all 3 of these cases, they are more of a nuisance then anything else.

Link to comment
Share on other sites

The ICT was done on the same donors twice not different donors 

and we do this two methods results comparison in our lab  and compare the results every 6 months and we picked up  on these donor this time 

the first donor case was resolved it was a cold antibody, I did a cold ICT using ortho surgiscreen on Reverse diluent cassette and patient sample with bliss no 37 incubation just room temprature incubtion and i was positive 

 

still resolving the second donor 

 

Link to comment
Share on other sites

20 hours ago, radwan1411 said:

Thats was in our laboratory policy as an indicator of quality control, for cap and aabb checklists, if there is a discripency as above there should be a corrective action after reaching to the result of the discripency 

Method comparison - apples and oranges in Blood Bank, but it makes CAP happy that you can check the box.

Link to comment
Share on other sites

We started doing this every 6 months for CAP years back.  I now just assign 2 pos and 2 neg samples from our proficiency samples (AFTER the due date of course) and techs will perform in gel, PeG and albumin.  Reaction strengths vary of course but it keeps techs comfortable with the tube methods which we rarely use.

Link to comment
Share on other sites

  • 4 weeks later...

For your A-ive donor with the screen positive in gel and negative in tube, is there a pattern in the antigram? Gel is much more sensitive to Rh antibodies than tube. You said the patient is male so unlikely they received RhIg. If results repeat in gel, I would suggest Rh phenotyping the donor for c, e status.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.