  1. I've experienced remote alarms that were monitored by facilities crew.  Even though the lab was 24/7.  Facilities even did the alarm checks.  Seemed to work pretty well though I had to tweak that system while I was their temp manager.  Alarm probe in freezer in the air - they wanted it to be sensitive, well it was.  The chart looked like a supernova explosion.  I told the medical director if I was inspecting they would be tossing everything out.  Once we put the probe in 50% glycerol the system worked pretty well.  I still did weekly checks on the documented temps for both refriges and freezer.  Otherwise, I agree, if you are 24/7 there is no need for a remote alarm.

  2. On 5/31/2022 at 7:29 AM, OneMore said:

     . . . inspectors don't like typed initials as 'proof' an individual ran the testing (especially if it's something that can be edited).

    How can you prove anyone did anything?  Unless you watch.  Inspectors cannot impune your work is bogus based on your process, unless something seems amiss.  I would immediately contact their regulatory agency and ask for the official stance on such and/or a replacement inspector. 

  3. Usually you don't have to revalidate after a scheduled downtime unless you had some changes made, especially if it involves any automation you are using.

    However, if you have an unscheduled downtime you will need a validation plan - to make certain your truth tables are still intact.   I've had these in the past but I no longer have a BBIS .

  4. On 5/17/2022 at 9:20 AM, John C. Staley said:

    Just a thought but, might that not be a little problematic if there is an issue with the reagents? 

     

    :coffeecup:

    I've never had a problem w BB reagents not working (except when I set them up to fail).  I have seen problems with reactivity not on par with usual results.  Contacting vendor discovered a change in the formula at production. 

  5. 58 minutes ago, exlimey said:

    Thank you, David. As I suspected. So why do two test systems using the same lot numbers of reagents need to be subjected to daily QC ? Let the debate begin.....:)

    When you say assay, that is, to me, a test, like Type and Screen.   Yes, we are using the same reagents but they are being used differently (automation vs manual).  Your QC in this instance validates the entire test system. 



  6. On 4/26/2022 at 11:15 AM, NancyC said:

    QC on Vision is sufficient. If all reagents and card lots are the same, the only difference is if the analyzer pipettes or a tech, it is all the same methodology. I verified this with CAP. We have had a Vision and manual gel since 2016, running QC on Vision only. We do also use the tube method and QC that separately as it is a different methodology. Curious as to which regulatory agency states Gel methodology is different if manually pipetted?

    Here is my response from CAP today, 4/28/2022:

     

    Good afternoon David;

    Thank you for contacting the CAP. Each test platform/system must have QC performed each day of testing. Regardless of the lots of reagents, the two systems you mentioned are separate. When I worked in the Blood Banks, we used the same Immucor reagents for the Echo instrument and the bench. We still performed the QC for each system. If you have any further questions, please feel free to reach out. Happy Lab Week!

    Linda Whaley, MLS (ASCP)
    Take a tour of our accreditation resources!

  7. On 4/26/2022 at 11:15 AM, NancyC said:

    QC on Vision is sufficient. If all reagents and card lots are the same, the only difference is if the analyzer pipettes or a tech, it is all the same methodology. I verified this with CAP. We have had a Vision and manual gel since 2016, running QC on Vision only. We do also use the tube method and QC that separately as it is a different methodology. Curious as to which regulatory agency states Gel methodology is different if manually pipetted?

    JCAHO in California.  I agree with this only because the manual QC not only qcs you gel cards but your manual pipets, manual pipetting and your mts centrifuge.  It is a different system.  However, if CAP says what you are doing is fine you should have that documented because if I was inspecting you I would cite.

     

     

  8. Decades ago that's how we used to store plts.  Good for 3 days at that time.  My understanding of such is that refrigerated plts are able to function upon transfusion whereas the current room temp plts require an "adjustment" period before they are clinically functional.   I'd like to hear the current feelings on such (just so I'm not laboring under a delusion).

  9. 1 hour ago, AMcCord said:

    My facility has never been cited by JC for therapeutic phlebotomies. Interesting. I handed that service over (quite happily) to the outpatient infusion clinic years ago. I did train them initially and give them copies of my forms as a documentation suggestion. My CAP checklist gets N/A for those questions and I've never been approached by a JC inspector. I would hope that they have yearly competencies, but I don't know for a fact. Might have to inquire.

    Me too.  My patients were/are rather unhappy w the Nursing approach to TP.  I'm not JC so I've never had that scrutiny. 

×
×