Human versus monoclonal reagents

Recommended Posts

Posted (edited)

Silly question...,but I'd really like to know,what is the difference in choosing between human and monoclonal reagents for ABO/D testing? Is it a first choice from these two?Is it the price?
I work with Bio-Rad reagents and I know they have cards with human reagents but also identical cards with monoclonal reagents.Our lab use only monoclonal ones.
 

Edited by Amra23

Share this post


Link to post
Share on other sites

A few things as far as human reagents.

Firstly, you never know what else may be in them in terms of antibodies directed against low prevalence antigens, because there is absolutely no way that the producer has the ability to test for all such specificities (I can remember once a human-derived anti-D reagent produced at one of the places I worked, also had a Gm antibody in it that we didn't know about.  It is highly unlikely that this would have caused too many problems, but there is, nevertheless, a small chance that this could have caused a false positive).

Secondly, you never know what else may be in them in terms of viruses, some of which may, as yet, be unknown to us (remember, HIV, used not to be known).  This is a danger to the producer and the person using the reagent, rather than the patient.

Thirdly, the avidity of human reagents is, in general, pretty poor (particularly anti-D).

A few things concerning monoclonal reagents.

Some of them cross-react with other specificities (although not many), but, famously, monoclonal anti-D reagents will react with the I and i antigens if used straight from the fridge.

They have to be blended by experts to ensure that the desired epitopes are detected, but certain Partial D types (e.g. Partial D Type VI) are not detected (unless required).

They are very specific and very avid (both of which are greatly to be desired).

Virally, they are almost certainly sterile.

Hope that helps.

Share this post


Link to post
Share on other sites
Quote

Thank you,Malcom!

''monoclonal anti-D reagents will react with the I and i antigens if used straight from the fridge.''

I didn't know that! This is something I will always have to remember!

Share this post


Link to post
Share on other sites
1 hour ago, Amra23 said:

I didn't know that! This is something I will always have to remember!

It is, nevertheless, true.  

Thorpe SJ, Boult CE, Stevenson FK, Scott ML, Sutherland J, Spellerberg MB, Natvig JB, Thompson KM.  Cold agglutinin activity is common among human monoclonal IgM Rh system antibodies using the V4-34 heavy chain variable gene segment.  Transfusion 1997; 37: 1111-1116.

Thorpe SJ, Ball C, Fox B, Thompson KM, Thorpe R, Bristow A.  Anti-D and anti-i activities are inseparable in V4-34-encoded monoclonal  anti-D: the same framework 1 residues are required for both activities.  Transfusion 2008; 48: 930-940.

Share this post


Link to post
Share on other sites

Also, the polyclonal (human) reagents will give false pos results in samples with pos DATS.  But anyway, finding sufficiently good human antibodies to manufacture reagents from is getting harder and harder.  So definitely monoclonal

Share this post


Link to post
Share on other sites
2 hours ago, Amra23 said:

I wonder why they still manufacture human reagents if that's the case...

Because, believe it or not, there are some dinosaurs out there who still demand them, because they don't trust monoclonal antibodies, and they are senior enough (God knows how) that they cannot be over-ruled.  If sufficient people still want them, people will still produce them and sell them to the dinosaurs, at higher and higher prices.

Share this post


Link to post
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now

  • Similar Content

    • By JCarpenter
      We are in the process of choosing new Coag. analyzers and have narrowed it down to IL and Siemens.  I was hoping to get opinions on the liquid protime reagent for the Top 350s that IL says is available.  Also how is their Heparin Xa reagent.  We have a CA1500 now and would change to the CA2500 but I have been having some QC issues lately but my instrument is only 12 years old.
    • By F Kloren
      In South Africa it is legislated that the ABO/Rh be confirmed on the segment from the pack before transfusion. When the blood bank changed to an automated crossmatch method, segment testing remained manual due to the cost of automated grouping. The cost of testing the ABO/Rh on the Olympus PK7300 proved to be significantly lower than other methods like Autovue etc.

      While there were substantial benefits new problems emerged. The tediousness of the task of cutting the segments manually to decant the blood into the test tubes (800-1000 per staff member), as well as manually comparing the bar-code numbers.
      We have now developed an automated segment cutter. The segment cutter prints the bar-code on the test tube, cuts the segments, decants the blood and adds saline. The racks with test tubes can now be placed directly onto the Olympus for analyzing. The decanter can process up to 300 segments per hour and sends the information to the LIS.
      Now my questions are the following:

      •    Are the cross matching processes the same globally, i.e. cross match on segment not on specimen?
      •    How common is it that the blood banks do all testing in bulk and send final product to hospital?

      I apologize in advance if I have some of the terminology wrong. I designed the machine and my blood testing knowledge is limited.
      Any information would be greatly appreciated.
      Thanks in advance.
       

    • By txlabguy82
      Hi All,
       
      So I would like to present a scenario that happened to me and get your input.
       
      I received a specimen from the ED for ABO/Rh testing on a young female (she had a miscarriage, which at the time I was not aware).  We use the BD Pink (EDTA) blood bank tubes for all of our blood bank testing, this particular sample was about a little more than 1/4 of the way full (yes not the best sample, learned my lesson with this case) - there were no visible clots in the tube or in the cell suspension I made for testing (testing was done fairly quickly since it was only an ABO/Rh and we use a STATSpin centrifuge so I had results out within 20mins of it being collected). 
       
      These were my initial reactions at time of testing:
       
      Anti-A:         0
      Anti-B:         4+
      Anti-D:       w1+
      D Control:    0
      A1 Cells:      4+
      B Cells:        0
      Du:               3+
      Du Control:  0     CCC: 3+
       
      So of course my interpretation was B Positive, which was reported. (We use the monoclonal Anti-D)
       
      All of our samples are retested by another technologist if we have no previous history on the patient.  The samples sit at room temperature until they are tested and this was one was retested roughly 10 hours after my initial testing, these are the results:
       
      Anti-A:         0
      Anti-B:         4+
      Anti-D:          0
      D Control:    0
      A1 Cells:      4+
      B Cells:        0
      Du:               0    CCC:  3+
      Du Control:  0     CCC: 3+
       
      Interpretation:  B Negative
      It was tested 2 more times by two other techs later that morning and the report corrected and the patient had to be called back to get Rhogam injection. So of course and event report was initiated at that point for root cause analysis.  
       
      I was approached 3 days later after knowing nothing about what happened and told that I "miss-typed" a patient sample.  After reviewing the work card I of course said No I didn't because I actually remember working on this particular sample due to the fact that the D got stronger at AHG phase.  I was extremely puzzled by the results and pulled the sample at this point had been in the refrigerator for 3 days and found that the same had numerous large clots and lots of visible small clots in my cell suspension (however none of the previous techs noted or expressed this to my director).  
       
      I performed (3 days later) a forward and reverse typing (B Neg), DAT, and IAT.  The DAT (IgG and Poly) - Both negative (very sticky microscopically) and IAT in Gel was completely negative.  So at this point it was very perplexing as to what happened with the Anti-D.  Of course everyone my boss talked to said that this was impossible for the Anti-D to just disappear (I was not inferring that it disappeared to her) and I must have done something wrong, which really aggravates me to use the word "impossible" in medicine.  She said it was more logical that I mixed the control tube and the Anti-D tube at the AHG phase when I read the tubes which makes absolutely no sense to me (if the results were reversed from the results I got at immediate spin, then yes that would make sense and I would have questioned the results and started over).   
       
      OK, so here is my theory as to be the possible cause for this particular scenario has to do with a sub-optimal sample that was in the process of clotting at the time of initial testing - however I can only assume this theory and I know that it didn't interfere with the Anti-A or Control on the forward type, but these are all different antiseras with different blends of antibodies/proteins, etc.. I am thinking that since I didn't see or detect clots when I tested the EDTA sample the first time it is possible that something during the clotting process potentially interfered with the Anti-D causing it to be a false positive. Since the specimen wasn't tested again until 10 hours later when the clotting process was definitely complete there is no way to prove this, unless someone where to have retested it immediately after I performed the first test.  We've seen cold autoantibodies disappear after sitting at room temperature due to autoadsorption, who is to say that since the clotting process was complete whatever was causing the interference may have been gone 10 hours later when the full clots formed.   My boss refuses to even think this is remotely a possible and I had to have made an error, she said we use to use plain red clotted tubes all the time for blood bank and never had any problems (don't think she is getting that these tubes have no additive and normally the clotting process was a lot faster in plain red tubes with no additive. Then again she also couldn't understand how a clot would get stronger overtime (I wanted to bang my head on the desk with that remark).  If I made an error or potentially made an error I would definitely own up to it, but in this instance her suggestion that I switched the test tubes around and reading the control as the patient makes no sense what so ever. 
       
      Has anyone every encountered anything like this before and what is your thoughts on the potential reason for the various reactions between a 10 hour period?
       
      Thanks,
      -TxLabGuy82
    • By bldbnkr
      Hi everyone,
      I just received a notification from Ortho about recently transfused patients and interpretation guide.
       
      "Following centrifugation of a freshly collected patient sample drawn from a recently transfused patient, the potential exists for the donor's transfused red cells, which are denser and heavier, to concentrate at the bottom of the sample tube in a layer below the patient's own red cells, which are less dense and lighter.  If the cell suspension used for testing contains a majority of transfused donor cells, unexpected patient results could be observed due to the patient cells being an absent or minor population in the prepared cell suspension.  The unexpected patient result could be a result associated with the transfused donor cells, or it could be due to a mix of patient and donor cell populations"
       
       
      Currently I use gel for antibody screening and id and gel crossmatches only...no antigen typing or ABO typing. 
       
       
      Do you think that I would have to change anything in my SOP's?
       
      Thank you.
    • By kirkaw
      Greetings,
      We use the Ortho Confidence system to QC our wet reagents. I have been alerted that there will be a delay in the next lot number shipment due to the (US) government shutdown. The current lot that we are using expires 10/29/2013 and the new lot # is set to ship on 10/30/2013.
       
      My question is, can I use patient samples that have established values for QC testing for the couple of days that I am without unexpired QC material? If so, how many examples of a particular value do I have to have in order for it to be a valid QC evaluation? For instance, if I want to use a group A patient to QC the anti-A reagent, how many times does that patients blood group have to have been tested?
       
      Thanks for the input.