Room Temperature antibody positive cell


Recommended Posts

Hi everybody

I hope you are all well

If I found one cell of the antibody screening cells positive only for room termperature phase only should I do antibody identification or not and what is to be written in the report and is it important in the crossmatch

Thanks

Link to comment
Share on other sites

From my point of view, I would say, no, nothing and no!

I can't believe I have a slightly different view from Malcolm......(Please don't beat me up, Malcolm!)..

Not knowing any details about what type of testing techique eltonsey is using, etc., I would be a little more cautious. I have to admit, I would do an Antibody Panel to see what I am dealing with. Years ago (before our current enhancement solutions, techniques, and automation), I saw a few patients who were just starting to produce new antibodies that were most IgM (and little or no IgG yet). Those antibodies might not have caused serious transfusion reaction if the patients were transfused, but they could be important information if a patient was a pregnant woman.

Donna

Link to comment
Share on other sites

Are you performing any sort of incubation prior to centrifugation of the card? The testing performed in IgG cards is listed as an AHG process according to the Technical manual and in the MTS instructions for use of the IgG cards. I don't think the process is considered a room temp procedure, although I do follow your logic I think.

Link to comment
Share on other sites

Agree with Mr. Needs.

The only reason to do RT is with the xm is to look for ABO incompatibility (If you are doing xm at all). RT incubation antibody screens are not for routine testing. Only for complex antibody workups and Reference labs.

Link to comment
Share on other sites

Actually, as a Reference Service manager, I can assure you that we wouldn't do room temperature tests in the UK, unless it is likely to help with a complex serological case. Usually, they are a complete waste of time!

Agreed, as a former Reference Lab manager we used them only when trying to reproduce what our clients were seing so could help them understand what was happening. Why it is still hanging around in the US hospital transfusion services is mystery to me.

Link to comment
Share on other sites

Dear Eltonsey. Thank you for that precision. In that case, the correct method, which I am sure is what you are using, is to pipette 50ul of the I-II-III cells followed by 25ul patient plasma. Incubate 15 minutes at 37°C, then centrifuge and read. If you had 1 cell positive in this method, with a negative panel, then the ptient maybe had an antibody against a low frequency antigen. In that case, if you let me know the lot number of the I-II-III cells you were using, I can see if maybe I can get any more information. In this case, if you do a normal crossmatch, you should be OK. On the other hand, there used to be a method where 50ul of plasma is used instead of 25, and the LISS Coombs cards are incubated for 10minutes at room temperature. Although this method still exists, it is not to be recommended for routine use - but sometimes can be quite helpful in elucidating funny anti-S or M or N antibodies. If this is the method you are using, I recommend you change immediately for the normal method (ie 15 mins at 37°). If you need any other help, please let me know

Anna

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.