We were just given a deficiency from CAP for having our Rhogam stored in fridge 1-8, supposed to be 2-8 . How do you all alleviate this ? or proove your fridge never drops to 1 degree? 

checklist all Common 30350

Thank you so much

  • Replies 7
  • Views 3.2k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • David Saikin
    David Saikin

    You have to change your refrigerator alarm settings to activate at 2.2C (or whatever you choose).  It doesn't matter that your refrigerator never gets too cold;  if you are going to store reagents (an

  • Even if you don't store RhIG in BB, you need to check the storage requirements of your BB reagents. We assume all reagents are 1-8C or 1-10C but there are some that are 2-8C. We borrowed some Anti-E f

  • David Saikin
    David Saikin

    You change the alarm setting on your refrigerator to alarm just before 2C (2.2 - or whatever you are comfortable with).  It doesn't make a difference if you say your refrig never gets to 1C.  Even if

comment_82000

You have to change your refrigerator alarm settings to activate at 2.2C (or whatever you choose).  It doesn't matter that your refrigerator never gets too cold;  if you are going to store reagents (and RhIg) in a BB refrig, it has to be able to alert you when then temp is out of prescribed range.

comment_82047

Even if you don't store RhIG in BB, you need to check the storage requirements of your BB reagents. We assume all reagents are 1-8C or 1-10C but there are some that are 2-8C. We borrowed some Anti-E from a local hospital that was from a different manufacturer  that we use. I happened to look at the package insert just before our recent CAP inspection. I had to adjust the fridge until the reagent was used up. Just be aware.

 

comment_82058

We use a certified alarmed thermometer that will alarm, separate from the refrigerator alarm and the wireless, set for 2.5 to 5.5. This is checked during quarterly alarm testing.    

  • 1 month later...
comment_82144

Change your setpoints to 2.5-5.5 - that way you are covered for both RhIG and blood products

comment_82145
On 5/21/2021 at 1:38 PM, DebbieL said:

We borrowed some Anti-E from a local hospital that was from a different manufacturer  that we use.

Did you also validate the anti-E to ensure that it was as effective (or better) than the one you usually use?  We would have had to have done this in the UK (not joking).

comment_82148
On 5/13/2021 at 8:08 AM, Theresa Garcia said:

We were just given a deficiency from CAP for having our Rhogam stored in fridge 1-8, supposed to be 2-8 . How do you all alleviate this ? or proove your fridge never drops to 1 degree? 

checklist all Common 30350

Thank you so much

You change the alarm setting on your refrigerator to alarm just before 2C (2.2 - or whatever you are comfortable with).  It doesn't make a difference if you say your refrig never gets to 1C.  Even if you have the charts to prove it.  If you cannot be warned of an impending out-of-compliance temp you are not in compliance.