Storage of Rhogam - Deficiency !?


Recommended Posts

We were just given a deficiency from CAP for having our Rhogam stored in fridge 1-8, supposed to be 2-8 . How do you all alleviate this ? or proove your fridge never drops to 1 degree? 

checklist all Common 30350

Thank you so much

Link to post
Share on other sites


You have to change your refrigerator alarm settings to activate at 2.2C (or whatever you choose).  It doesn't matter that your refrigerator never gets too cold;  if you are going to store reagents (and RhIg) in a BB refrig, it has to be able to alert you when then temp is out of prescribed range.

Link to post
Share on other sites

Even if you don't store RhIG in BB, you need to check the storage requirements of your BB reagents. We assume all reagents are 1-8C or 1-10C but there are some that are 2-8C. We borrowed some Anti-E from a local hospital that was from a different manufacturer  that we use. I happened to look at the package insert just before our recent CAP inspection. I had to adjust the fridge until the reagent was used up. Just be aware.

 

Link to post
Share on other sites

We use a certified alarmed thermometer that will alarm, separate from the refrigerator alarm and the wireless, set for 2.5 to 5.5. This is checked during quarterly alarm testing.    

Link to post
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.