Platelet Incubator on the Fritz


Everett9239

Recommended Posts

I have a question that we are having trouble wrapping our minds around.  Our platelet incubator is getting too warm, so we have been using it with the door open.  Because we are doing this, we have been taking the room temperature every four hours.  Do we need to continue doing this?  The platelet incubator chart recorder is still fully functional and the alarm is still functioning as well.

Link to comment
Share on other sites

I think you would have to validate the chart recorder and alarm for the new way (door open) that you are using the equipment. But that means that you are accepting that  the door hanging open is 'normal' operation, which of course you are not. I would take the temp every 4 hours or get an external chart recorder/alarm if it's going to take awhile to get things fixed.

Link to comment
Share on other sites

I'm afraid I fall on the "Why bother" side of this discussion.  All you care about is the temp in side the incubator.  If the door is open and the chart indicates the temp is where it should be I think you are just fine unless there is someting in the operators manual that indicates that the temp sensors and chart fails to function when the door is open.  Is your department staffed 24/7?  If so, I fail to see a problem.  But then I've had more than a few issues with inspectors over the years.  :bonk:

Link to comment
Share on other sites

We use to have this problem quite a lot- and realised that the heat exchangers were getting iced up and the temp started rising. So now we defrost every 2 months as part of the deep clean schedule ( of course this was after getting the engineer out to fix quite a few times... and as it had been turned off overnight, everything was always fine when he arrived- just cost us ££££ !)

Link to comment
Share on other sites

I'm afraid I fall on the "Why bother" side of this discussion.  All you care about is the temp in side the incubator.  If the door is open and the chart indicates the temp is where it should be I think you are just fine unless there is someting in the operators manual that indicates that the temp sensors and chart fails to function when the door is open.  Is your department staffed 24/7?  If so, I fail to see a problem.  But then I've had more than a few issues with inspectors over the years.  :bonk:

I agree John, I don't see a problem either.   

Link to comment
Share on other sites

It seems like if your chamber is warming ABOVE room temperature, then you might be better off taking the platelet rocker out of the chamber, placing it on a counter at ambient temperature and monitoring the room temperature until you can get your replacement.

Link to comment
Share on other sites

The advantage I see to leaving the rotator in the chamber with the door open is you still have the functioning chart recorder and the alarm system to warn you if the temp gets out of range.  If you put the rotator on the counter you will need to document the temp every 4 hours.  I'm kinda lazy and prefer to let the equipment work for me.  :faint:

Link to comment
Share on other sites

OK, if you can't fix it I'm with John, leave the rotator in the box. Lucky guy, getting a new one. It's a sore subject for me since I've asked for an incubator for, oh, decades. Our ambient temperature flucuates, and we end up running around the lab to move the rotator to a warmer or cooler spot to keep the temp in range. Ridiculous.

Edited by Dr. Pepper
Link to comment
Share on other sites

If you had that external chart recorder/alarm, you wouldn't have to take temps every 4 hours AND it would alarm so you'd know when you have to move it to a cooler/warmer spot in the lab.  :clap: 

 

On a happier note, my incubator - 23 years old and starting to fade - is finally scheduled for replacement in the 3rd quarter this year. Yippee!

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.