Hallo!

Pt. with 0D CcD.ee K-

DAT 3+( IgG3+,IgA-,IgM-,C3c-,C3d-,ctl-)

plasma: 2+reactions with all e+ cells in panel

eluate:Auto-Anti-e (acid-elution with BAG)

needs RBCs,

we should do an auto-absorption to get the auto-anti-e out of the plasma before we use it for xmatch to rule out an underlaying alloantibody, right?

without absorption autocontrol is 2+, xmatched rbcs 2+

can anybody tell me a good (and simple,fast...) way to do an auto-absorption?

seperate the plasma, wash the pt.cells with warm 0,9% NaCl- and .....?? we don´t have any other elution system than BAG, so we can´t use eluated cells to do the absorption.

Thank you,

K.

Link to comment
Share on other sites

The patient has not been transfused in the past 3 months, right? Autoadsorption is more effective if you can use an enzyme mixture to remove the antibodies from the patient's cells and make them better at adsorbing antibody before you try to do the adsorption. If the antibody is strong, you often need enough patient cells to do multiple serial adsorptions and that quantity of cells is hard to get from a patient that has hemolytic anemia.

Link to comment
Share on other sites

The patient has not been transfused in the past 3 months, right? Autoadsorption is more effective if you can use an enzyme mixture to remove the antibodies from the patient's cells and make them better at adsorbing antibody before you try to do the adsorption. If the antibody is strong, you often need enough patient cells to do multiple serial adsorptions and that quantity of cells is hard to get from a patient that has hemolytic anemia.

A mixture of enzyme (papaine) and DTT (a chemical not a enzyme) works the best (according to Petz), it will ZAPP that antibodies from the cells.

Peter

Link to comment
Share on other sites

I am a big fan of the PeG (polyethylene glycol) autoadsorption technique. It is fast, efficient, cheap, and doesn't require any special reagents other than a routine PeG additive (we get ours from Immucor). The procedure (from the AABB Tech Manual) is attached.

Link to comment
Share on other sites

I am a big fan of the PeG (polyethylene glycol) autoadsorption technique. It is fast, efficient, cheap, and doesn't require any special reagents other than a routine PeG additive (we get ours from Immucor). The procedure (from the AABB Tech Manual) is attached.

We started using PEG autoadsorption this year and it has pretty much replaced doing ZZAP-adsorptions.

It generally works great!

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.