Heating Block / Incubator


BBK710

Recommended Posts

What kind of quality control do you perform on heating blocks or incubators used for tube testing other that daily temps? We have a procedure to determine incubation time by checking how long it takes for the reagents and cells in the tubes to heat to 37 degrees and then adding that time to the incubation time stated in the manufacturer's inserts for each procedure. Does anyone do this? I don't know where we got the procedure because it isn't in the current technical manual.

Link to comment
Share on other sites

In the mid 70's, I worked at a hospital that incubated the tubes in a 37C waterbath. There was a reference (don't remember) that you had to add time (I think it was 5 minutes) to the incubation if you used heat blocks instead of water baths because it took longer for tubes in incubators to reach temperature. So they incubated in a waterbath to save the 5 minutes. That may be where this procedure originated.

We also performed screens in duplicate, leaving one set at room temperature for 15 minutes; did a 15 minute room temperature saline crossmatch in addition to the AHG (polyspecific, of course) crossmatch; and performed minor crossmatches. Can't even imagine doing that now!

Obvously, I am too old to be doing this job!

Link to comment
Share on other sites

We did perform an initial validation to demontrate that all wells in the heatblock were at the appropriate temperature and that a couple of antibodies, when incubated for the prescribed length of time, showed the appropriate level of reactivity. After that - daily temperatures.

Link to comment
Share on other sites

In the mid 70's, I worked at a hospital that incubated the tubes in a 37C waterbath. There was a reference (don't remember) that you had to add time (I think it was 5 minutes) to the incubation if you used heat blocks instead of water baths because it took longer for tubes in incubators to reach temperature. So they incubated in a waterbath to save the 5 minutes. That may be where this procedure originated.

We also performed screens in duplicate, leaving one set at room temperature for 15 minutes; did a 15 minute room temperature saline crossmatch in addition to the AHG (polyspecific, of course) crossmatch; and performed minor crossmatches. Can't even imagine doing that now!

Obvously, I am too old to be doing this job!

Oh - those were the days! :highfive:

Link to comment
Share on other sites

Validation (or re-qualification if after a move etc) and then it is monitored on the REES Centron temp monitoring system.

If temp stays the same, time to reach temp stays the same for the same type of tube used in the Validation proces - change tubes and you will have to re-do operational qualification only.

Cheers

Eoin

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.