Any Ideas Greatly Appreciated


David Saikin

Recommended Posts

No Patient history. Types as group O, Rh(o)D Positive; DAT: negative

Specimen was referred due to hemolysis noted in gel testing.

Gel panel displayed some weak mf rxs consistent with a cold agglutinin. Ficin pretreatment removed this reactivity A qucik cold panel in tubes revealed a cold auto at 4C with 2 O cord cells reacting as strong as the R1 and R2 cells (2+) - auto and rr were 1+.

A tube panel run with serum/LISS displayed mild hemolysis in all tubes (including auto) after 37C incubation. These cells were reacted with anti-C3b,-C3d after washing - all tubes 4+. A PeG panel with anti-IgG was pristinely negative.

I am open to any and all suggestions. Thanks.

Link to comment
Share on other sites

Enzyme treated cells, in tubes with serume and LISS - no hemolysis at 37 and no agglutination at AHG with anti-C3b,-C3d. I am calling it anti-Pr (if only I had some dog blood . . .).

David , what is the use of dog blood? Thank you!

Link to comment
Share on other sites

No Patient history. Types as group O, Rh(o)D Positive; DAT: negative

Specimen was referred due to hemolysis noted in gel testing.

I am open to any and all suggestions. Thanks.

Gel 0.8 cells often show hemolysis when they are contaminated. That is where I go first. Light and being out at room temp too long. I instituted rotating racks so no rack was out more than 8 hours and we keep the 0.8% cells inside a black sleeve to protect from light.

This is frequently the case and often on older cells. When we get a new set the hemolysis goes away.

Link to comment
Share on other sites

Interesting - I never see hemolysis with my 0.8% cells, even in panels which are 3 months outdated (I always wonder what Ortho does to accomplish that -AND if their treatment accounts for some of the troublesome reactions we see).

Gel 0.8 cells often show hemolysis when they are contaminated. That is where I go first. Light and being out at room temp too long. I instituted rotating racks so no rack was out more than 8 hours and we keep the 0.8% cells inside a black sleeve to protect from light.

This is frequently the case and often on older cells. When we get a new set the hemolysis goes away.

Link to comment
Share on other sites

The vials don't look hemolyzed. It's the reaction in the gel which looks hemolyzed. I don't see problems with panels either but they are kept in the fridge. Not routinely sitting on the counter for an entire shift or more. I asked ortho and they said that the sets are validated for 8 hrs at RT. They ASSUME them to be going in and out of fridge. But as a small but busy hospital and with reagents needing to be RT for use we found in our facility that was not happening. They were frequently out 2 shifts st least.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.