Computer Crossmatch: what is needed to implement?


Barbarakym

Recommended Posts

You BBIS must be (501k) approved for the e-xm. (truth tables for pt/donor types, written policy - need 2 valid ABORh types on pt - 1 may be historical). Validate that you cannot do this with only one type and incompatible ABO and that it does work when your acceptance criteria is met (2 pt types, ABO compatible).

There must be more.

Link to comment
Share on other sites

We are getting new system to us, Cerner. Supposedly this will be validated company wide. Some hospitals are up live already Can we assume the validation by company is good enough?

How about from our BB assuming computer as built is ok, what needs to be done local to switch from tube to EC? A procedure of how. I get that. Training also. But do we need to do anything else?

Switching from tube abs for example to gel we had to validate and each tech had to demonstrate. What do we need to do to validate it works as a whole ? Or cuz the company says it works we can just start handing out blood?

Link to comment
Share on other sites

We had a company wide validation when we went on Cerner ME, but you still have to validate the computer crossmatch works in your facility using your policies. There were some options in the electronic crossmatch program that needed changing before we went live that passed the company validation. There is one option where you can't add on units using the electronic crossmatch after issuing a unit without an override. That had to be changed. We use a Blood Bank bracelet and they didn't have a spot in the electronic crossmatch to enter the bracelet number so it would print on the tag. We resolved those problems and then validated in-house. The techs were given an in-service and had to pass a written quiz. We waited a couple of months just to get everyone used to the new computer before changing to the electronic crossmatch. Some of the facilities went live without validating in-house and were calling me to see why they couldn't use the electronic crossmatch after issuing a unit.

Link to comment
Share on other sites

Thank you everyone. Another related question. Once we go to computer (electronic) crossmatch are we required to be inspected by FDA?

No, just keep in mind, just because you are not collecting or modifying blood products, which would require inspections with the FDA, does not mean they can pop in anytime to inspect any Blood Bank. Just because they can. :)

Link to comment
Share on other sites

Do you mean that in the UK, electronic xm is not allowed if blood typing is done manually? That is not a requirement in the US.
Mabel, my understanding of the UK guidelines is that electronic issue (EI) is acceptable in the UK on manual typing, however it is strongly discouraged. It is unlikely that any laboratory in the UK would undertake EI without at least one group and screen being processed through an automated analyser with the results being transferred directly to the blood bank computer system following authorisation on the analyser.

Steve

:):):)

Link to comment
Share on other sites

Yeah it is strongly discouraged to the point they asked about it in the MHRA inspection... You can see why - a closed system is best. Barcode from wrist to wrist is what they hope for.
I entirely agree with the point you have made. We can discuss the merits (or otherwise) of testing two samples or one sample twice etc. The focus must be on the whole process from sample collection to transfusion (wrist to wrist) including the use of printed barcoded PID labels on the samples provided they are generated at the bedside. The less human/manual manipulation the better. We can dream!!!!!

Steve

:):):)

Edited by Steven Jeff
format
Link to comment
Share on other sites

Barbarakym,

Although venders must validate their systems prior to gaining 510k clearance, additional testing must be performed to include site specific build and design. Cerner and other companies offer third party blood bank validation services that meet the FDA requirements. Most of this work is done remotely and the only instance that I’m aware of where additional testing must be performed on site is for Computer Crossmatch. This is due to AABB standard 5.15.2.1 (current edition).

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.