DAT's


Recommended Posts

Hello All!

We currently do DAT's in tubes, Polyspecific first, and then if it is Positive, we will do IgG and C3d separately.

I would like to switch to doing DAT's in gel, because of the increased sensitivity (and we just got 3 ProVues). This presents a problem, however, because Ortho does not have a C3d gel card.

How many people out there are just doing IgG and Polyspecific only and letting the physician figure that if Poly is Positive and IgG is negative, it's probably due to Complement??? This is what I want to do, so we don't have to keep d oing the C3d portion in tubes, but I can't find any documentation in the Tech Manual to support this.

Any recommendations???:confused:

Thanks

Denise

Link to comment
Share on other sites

Hi,

We started doing DAT in gel almost a year ago.

We call the gel portion of the test a sensitive "DAT screen". The gel is a combination of the IgG and c3.

If negative the test stops there.

The positives reflex to both the IgG portion in gel and the C3 test in tube.

We result as either positive for the complement portion, positive for the IgG portion, positive for both or negative.

The docs are suppose to understand that the screening is more sensitive and a final negative is not out of the question.

Link to comment
Share on other sites

Hello All!

We currently do DAT's in tubes, Polyspecific first, and then if it is Positive, we will do IgG and C3d separately.

I would like to switch to doing DAT's in gel, because of the increased sensitivity (and we just got 3 ProVues). This presents a problem, however, because Ortho does not have a C3d gel card.

How many people out there are just doing IgG and Polyspecific only and letting the physician figure that if Poly is Positive and IgG is negative, it's probably due to Complement??? This is what I want to do, so we don't have to keep d oing the C3d portion in tubes, but I can't find any documentation in the Tech Manual to support this.

Any recommendations???:confused:

Thanks

Denise

Ortho BioVue do have Anti-IgG/Anti-C3b, C3d/Control cards (at least in Europe). We routinely use them.
Link to comment
Share on other sites

I assume the c3d check cells are run in gel in a separate well from the patient test. In tube testing we run check cells to make sure washing was adequate; since there is no washing in gel, it seems to me we should only have to run the check cells once each day of use, to prove the anti-c3d reagent is active.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.