With the PR platelets we are irradiating less. But I've found that the IPDM will still let you scan a PR platelet Ecode without any sort of warning. I've asked the vendor for direction, but wondering if anyone here has had a way of warning techs.

  • Replies 4
  • Views 1.5k
  • Created
  • Last Reply

Top Posters In This Topic

comment_82296
On 7/21/2021 at 12:55 PM, David Saikin said:

IPDM?

I Googled it.

Irradiated Products Database Management (IPDM)

comment_82297
On 7/20/2021 at 11:46 AM, rosi0017 said:

With the PR platelets we are irradiating less. But I've found that the IPDM will still let you scan a PR platelet Ecode without any sort of warning. I've asked the vendor for direction, but wondering if anyone here has had a way of warning techs.

 

I am a little confused; as the process to manufacture PR platelets has the same effect as irradiation what do you want to warn your techs of?   

comment_82305
On 7/20/2021 at 10:46 AM, rosi0017 said:

With the PR platelets we are irradiating less. But I've found that the IPDM will still let you scan a PR platelet Ecode without any sort of warning. I've asked the vendor for direction, but wondering if anyone here has had a way of warning techs.

I have it built into the computer system.  It won't let you irradiate a PR platelet.