red cell button sticking to tube


KPayne

Recommended Posts

We are observing that red cell button seems to stick to 12x75 tube when reading reaction with A and B antisera. Have looked at tubes,centrifugation, saline, individual techs, mixing of EDTA sample - would appreciate any suggestions to remedy this problem.

Link to comment
Share on other sites

We have had a similar problem with shaking off Anti-A. I think it's too long a spin time. Our current spin time is 20 seconds as determined by the centrifuge calibration procedure done once a year. I think it should only be 15 seconds (which is the default for the immunofuge), but what do you do if the calibration procedure tells you otherwise?

I told the tech who performs the calibration procedure to find a way to show that 15 seconds works, but the procedure of diluting the antibody used for testing is not real life situations. We have been using patient samples for the anti-A or anti-B, since diluting the monoclonal reagents requires a 512 or 1024 dilution to get that 1+ reaction the calibration procedure wants. She always gets 20 seconds as the ideal spin time!!

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.