Helmer waterbath temperature


Eagle Eye

Recommended Posts

Hi Donna,

I am curious about your last statement - "And we document in our computer system what the water temperature was during the thawing process." Do you use a thawer that alarms when it reaches unacceptable temps?

JB

We use a Helmer 4-unit model that does alarm when it reaches unacceptable temps. However, when we change (ie: "process") our frozen product into a thawed product in our BB computer (HCLL), there is a required field where we must document the temperature of the thawing bath (obviously, for QC documentation.) (Is this what you were asking, Joan?)

Donna

Link to comment
Share on other sites

Do you perform quarterly checks on both the high alarm and the low alarm? The manufacturer only indicates a high alarm check but if this is considered a blood product storage device (according to CAP inspector) then wouldn't we be required to do both?

Link to comment
Share on other sites

Do you perform quarterly checks on both the high alarm and the low alarm? The manufacturer only indicates a high alarm check but if this is considered a blood product storage device (according to CAP inspector) then wouldn't we be required to do both?

I'm having trouble understanding how a CAP inspector can consider a plasma thawer to be a storage device. Isn't that like calling a platelet centrifuge or an apheresis instrument a storage device?

Link to comment
Share on other sites

Recently we had a FDA inspection in which they focused on equipment preventive maintenance and quality control. They requested from us a list of equipment for the blood bank and reviewed our QC SOPs and also the manufacturer's Operator and Service Manuals. When the Operator's or Service Manual listed a recommended quality control or preventive maintenance the FDA looked for these and cited us for not having some of these recommended PMs in place, as well as, for not documented these recommendations. In the case of the Helmer Plasma Thawer, there is a preventive maintenance schedule indicating the frequency of performance with the statment by the manufacturer "Maintenance tasks must be completed according to the following schedule". For the "QuickThaw Plasma Thawing System", that we use, it requires weekly cleaning of the chamber and basket, quarterly lubrication of the moving parts, quarterly temperature calibration for the temperature controller readout, quarterly high temperature alarm and annual checking of the bearings for the basket. Also, we have our annual electrical check that is required.

They actually spent over 3 days just on equipment alone and we had to work very closely with our contracted Bio-Medical department to correct our deficiencies, as Bio-Medical department was documenting performance electronically and not very well (documenting the PMs and QC that they were responsile for was done, but not what the readings were or not documenting electronically who performed testing).

It was quite an experience, as there were some preventive maintenance recommendations that were impossible for my staff to perform. Some of the citations were resolved after phone calls to the manufacturer indicated that our original Service manual was outdated and the preventive maintenance shcedule was changed so that an anuual preventive maintenance performed by the manufacturer covered a particular preventive maintenance requirement (from semi-annual to annual). Also, although we had a cleaning schedule defined in some of our SOPs, we did not document these cleanings, so we had to develop a new form or revise forms for documenting these cleanings (if it wasn't documented, it wasn't done).

But I digress....

We have our high alarm going off at 36.7 degrees and our set point at 35.5, which in our case the actual temperature is a little higher (within 0.5 temperature allowable difference). We thaw both FFP and Cryo in this thawer. When I looked in the CFRs, temperatures of storage are not listed as 1.0 degrees or 6.0 degrees, but simply as 1 and 6 degrees. We all know that if we listed the storage temperature of blood as 6.1 degrees, we would be cited because it would be out of range. Using the same logic, if you have your high alarm point set up as 37.1 or higher, then these temperatures would be out of range and potentially you could effect the potency of the thawed product if it reached 37.1 degrees, which the FDA could construe as being a reportable error. Remember, it just has have the potential to effect the potency of the unit to be reportable. If someone would like to put this question to the FDA, be my guest.

Edited by conwaysbb
spelling or syntax errors
Link to comment
Share on other sites

Does anyone set and do a low alarm test on their plasma thawer? The argument for setting a low alarm [alarm sounds if the temp drops below 30C] is that regulations state frozen plasma components must be thawed between 30 and 37C.

Link to comment
Share on other sites

When we clean out the thawer, we observe the temperature at which the time readouts turn to zero. At this point you can lower the baskets and the thawer is ready to go. Ours is set for 34.1 on the low side.

:clap::clap:

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.