Posted December 11, 20177 yr comment_71829 Hi Everyone, Does anyone have a validation protocol for washing platelets in the Cobe 2991 Cell Processor? Any guidance is much appreciated!
December 15, 20204 yr comment_81339 Did you ever figure out the validation protocol for washing platelets? If yes, would you please share the information? I am in the process of validating. Thanks
December 16, 20204 yr comment_81342 All you need to do is document platelet recovery (should be at least 80%) and pH. Nothing else is required. No platelet function tests, etc. since these are not used with unwashed platelets. Would recommend washing in platelet additive solution rather than saline.Washed platelet buffy coat platelet additive manual.pdf