If you are suspecting multiple allo antibodies and when you run enzyme panel:

1) what if cell is positive for enzyme sensitive and enzyme resistant antibody? Eg. If you are suspecting anti-Jka and anti-Fyb and when you are evaluating enzyme panel and the cell which is a) Jka+ & Fyb+ ---if non reactive? B) Jka+ & Fyb+---what if reaction strength is same or increased?

Thanks

Link to comment
Share on other sites

In a) - I would expect the anti-Jka is ruled out; in B) I would say that anti-Jka is probable but that does not r/o anti-Fyb. However, if I also used cell B) to qc my ficin along with commercial anti-Fyb, then I would feel better that Fyb ag was destroyed if my qc worked . . . does that make sense? I recently had a similar scenario - anti-Fya,-Jsa. The Fya was very obvious . . . when ruling out other abs, the Fya=Jsa+ cell was 2+. The only other cells I had that were Jsa+ were also Fya+b=. I did a ficin pretreatment of the cells and ran them against the patient plasma and also commercial anti-Fya AND also untreated vs the commercial reagent. Both enzyme treated cells were 2+ with the patient serum and Negative with the anti-Fya. The untreated cells were 2+ with the commercial antibody.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.