I'm really interested in finding out the failure rate of your automation to interpret ABO/D types. It would be very helpful to know the approx % failures you are experiencing.

The equipment I am most interested in are:

Provue/ Diamed (Gel cards)

Galileo -microplate grouping

ECHO microplate strips

Ortho Autovue (Cards- glass beads)

Many thanks!

Edited by RR1
forgot to add a poll- never mind!
Link to comment
Share on other sites

We are seeing very few failures for ABORh on our ECHO. One thing we do notice, backtyping is frequently weaker on the ECHO than it would be in tubes, but even 1+ positives will interpret correctly. We occasionally see very small agglutinates in A or B backtyping wells that should be neagitve - the ECHO will fail these interpretations. I wonder if it is a cold or rouleaux that causes these or micro clots in the EDTA specimen? Overall, we are happy with the job being done on the ECHO.

Link to comment
Share on other sites

Thanks- I am currently extracting these figures from my Galileo & ECHO, but similar observations with the reverse group being seen. Our immucor specialist said we should possibly be centrifuging our EDTA samples for 10mins prior to testing- is this what everyone else is doing?

Link to comment
Share on other sites

We centrifuge our patient specimens for 5 minutes. (If we centrifuged them for 10 minutes our manual tube technique testing would be faster than the Echo!) Personally, I doubt that this is the solution to the problem.

Donna

Link to comment
Share on other sites

We definitely have problems with the Galileo missing D positive patients. We centrifuge our samples for 5 minutes, which was what Immucor recommended to us. I doubt the centrifugation time affects the D typing. I have not done an audit to determine how many get missed. I changed our policy to require that a second type be done manually for any patient who is Rh negative on the Galileo and has no history here.

Link to comment
Share on other sites

Results on automation may not always match manual tube testing for several reasons. Sensitivity of direct hemagglutination used on the Galileo & Echo for ABO/Rh is 1+ or better - a very weak reaction will be called negative. Some of our techs would call some of these reactions 2+.

Different anti-D reagents are used on automation than bench - anti-D's are now all monoclonal, but each reagent is specific for a different epitope of the D antigen. We are seeing many more weak D and partial Ds than previous reagents (human origin pools). I think there is much more variation in the D antigen than we thought.

The big question is what do you want call your patients - D pos or D neg? Think carefully, if you have an OB/Gyn population, it's important to ensure that D partials are called D neg to qualify for RhIG. There are no serologic reagents that will tell you this definitively. So study your reagent reactions, do some on-line research and pick an anti-D that is negative with most D partials. We are in process of changing manual tube reagent to match reagents used on our Galileo specifically for this purpose.

Bottom line, many of us now have 2 methodologies in our labs. We are bound to see variations in results. Just as Chemistry lab what they see between methods!

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.