JHH1999
Members - Bounced EmailContent Type
Store
Profiles
Forums
Blogs
Events
Frequently Asked Questions
Gallery
Downloads
Glossary
Links Directory
Questions
Jobs
Vendors
Everything posted by JHH1999
-
saline expiration date
JHH1999 replied to KLCarter's topic in Immunohematology Reference Laboratories
Typically expiration dates are established by the manufacturer. They perform stability testing for the duration of the assigned expiration date to support it with data. Data shows the product is capable of performing up until the claimed expiration date. It may continue to function they just do not have data to support it. This should be done to support any "reagent" made that is not at least qualified in some way each day of use. Assigning an expiration date based upon the shortest dated component is not very good science. The different ingredients could be compatible with each other or could have a negative impact to on another. From a pure scientific and quality aspect one would prepare a reagent and place it on a stability schedule and test it periodically for performance. This data is then used to support the use. Any assessor should accept this science. A simple analogy I like. If you made a cake today with milk that had a use by date of tomorrow does that mean the cake is not good after a day. Of course not, since it is now in a different form and could be stored in a different way.- 18 replies
-
- saline
- expiration date
-
(and 1 more)
Tagged with:
-
Basically some labs struggle with QC and the actual intent. They also may have issues with staff not detecting weak reactions, specifically in tubes. Manufacturer's in the US tend to make QC material so strong that it is less likely to be missed. The result is the manufacturer avoids product complaints and the labs do not have QC failures requiring investigation. However, the intent of QC is missed in my opinion. QC should ensure your test system is capable of detecting weak reactions. QC products manufactured where I work were developed (IgG coated control cells and Daily QC Kit) to detect if there is a weakness rather than just checking a box that QC has been done.
-
As a reagent manufacturer one of the requirements is to ensure products work as intended over the shelf life of the product when in the end users hands. This includes ensuring the product is able to withstand shipping. Manufacturers perform stability testing of these products to a greater extent than most end users would be cabable of repeating. This would include making sure red cells antigens are viable. Daily QC of red cell panels by the end user is not necessary in my opinion and a waste of time and resources.
-
CAP checklist TRM.31250 Reagent Expiration Dates
JHH1999 replied to AngelaNlincoln's topic in Accrediting Agencies
Using expired antisera and panel cells is and has been standard practice for years in my experience. It has become more sound with additional QC people are now doing. If ever cited for this I would push back. Provide scientific rationale and a risk evaluation for what you are doing. These are tools used to solve problems. Without them your hands are tied and patients may not receive compatible blood. A bigger risk. No matter what you call them (inspectors, investigators, assessors) they all have opinions and you may not be about to change their mind during the inspection. What they may cite you for during an inspection can be disputed in your response. The repsonse is typically reviewed by others. Some of these things are just driving up the cost of health care without improving actual patient care. -
I agree with Malcolm. DNA work can predict, but does not show if the antigen is expressed. To complicate things, does it matter regarding transfusion therapy. Do we provide "gene" compatible blood? This could decrease your donor pool. Additionally, it will be interesting to see how insurance companies will see paying for something more expensive when less expensive serological phenotyping is available and still meets patient needs. DNA work is interesting and can be a good tool for investigating the strange and unusual, but it may complicate things too much for routine testing and currently not licensed by FDA.
-
I am not familiar with UK regulations, but in the USA blood grouping reagent specifications are listed under 21 CFR 660 Subpart C. The other document available is older, but it is currently the only guidance I'm aware of that indicates which low incidence antibodies are required to be ruled out. Bg antibodies are not required, but we try our best to rule them out to prevent problems for customers. There has been the rare occassion where a polyclonal reagent does contain a low incidence contaminant that is not required to be ruled out. The availability of rare cells can be a problem and where does it end with all the lows known to exist. No enzyme testing is required. http://www.fda.gov/downloads/BiologicsBloodVaccines/GuidanceComplianceRegulatoryInformation/Guidances/Blood/UCM080926.pdf
-
I have been performing serological tube testing my entire career (over 20 years) and I don't recall seeing anyone covering tubes when centrifuging for typing, screens or panels. We still use old reliable Clay-Adams Sero-fuges as well as cellwasher 2s. Someone I guess could be concerned about creating aerosols, but I think the risk is extremely low. It may be just referring to spinning down samples at initial receipt (larger volumes). If you were to take this statment seriously and cover all tubes when centrifuging, I would say good luck when the covered tubes are placed in an automated cell washer like a cellwasher 2.
-
Antigen testing donor units
JHH1999 replied to Whitney's topic in Donor Services and Donor Recruitment
At this point I do not believe molecular testing is licensed. Molecular testing can be used for screening and investigating strange serological reactions. Since it is not licensed I do not think you can label the units. You would need to confirm the antigen typing with a licensed test method. -
If you are not detecting the Anti-E on the Tango and also now not detecting it in gel or tube with PeG, it does not sound like a problem with different methods.
-
Glad I could help. Now being in the reagent manufacturing world we get to see some strange things like this come up when trying to resolve customer problems with reagents. This just happened to be one we saw recently. Reminds me of old reference lab days.
-
This sounds like a little known phenotype that has been previously identified. The patient is most likely S-s+U-Dantu+. We found a donor with this phenotype not long ago. It was also described in Marion Reid's The Blood Group Antigen Facts Book.
-
Storing of DTT reagent
JHH1999 replied to emergency room's topic in Immunohematology Reference Laboratories
We have found DTT to be very stable when frozen. As long as you run a proper control to ensure the DTT is working each day of use, it should be fine for many years. It may be wise to establish some stability data (what manufacturers of reagents do) that can be used to support its use over time if you are making it up yourself. This can be done by preparing DTT and storing it frozen in aliquots. Then periodically test it to ensure it denatures a Kell system antigen for example. This data can then be used to support an assigned expiration date. -
Anyone Using Bio-Rad's Blood Bank Reagents?
JHH1999 replied to jhaig's topic in Transfusion Services
I am not aware of any regulatory requirement for blood banks to validate the use of a FDA licensed reagent. The FDA does not validate reagents. They review and approve the work done by the manufacturer to ensure their claims are accurate. If you follow the manufacturer's instructions and run the necessary controls daily, you should be able to put them into use without any further testing. Any additional work would be in my opinion a waste of time and valuable reagent. As far as the gel thing, I have heard something about the patent on that technology ending allowing others to start making gel system supplies to compete with Ortho. -
Anyone Using Bio-Rad's Blood Bank Reagents?
JHH1999 replied to jhaig's topic in Transfusion Services
Just an FYI. All licensed blood grouping reagents must meet FDA requirements for potency, reactivity and specificity. In order to license these reagents they must undergo significant clinical trial testing to ensure they are giving accurate results. It is one thing to compare or evaluate how these reagents perform against each other in your hands, but the manufacturers have performed far more testing than most end users would even think of doing. I would caution against going overboard on "validating" the use of these reagents. As previously mentioned, pay close attention to the directions for use. Some newer monoclonals have different incubation/spin time requirements and even different saline requirements. If these are not followed, your reactions may not be what you would expect. -
I would agree with this approach on this subject. Hence, why as stated it is very important to review and follow manufacturers instructions if you want to obtain valid results. Different manufacturers reagents may have different instructions based upon formulation more so now than ever with the arrival of so many monoclonal reagents instead of traditional polyclonals. However, the methodology of these reagents is still the same. We are using antibody to detect the presence or absence of antigen on a red cell. The antigen typing using one reagent should be the same using another reagent unless there is some rare type of variant present. Being in the blood bank world for 20 years now including the last 10 years in quality and validation of reagent manufacturing, I'd like to save folks what I see as unnecessary additional work and effort.
-
I'm not sure how anyone could validate a blood grouping reagent from an end user standpoint or what would be the reasoning to perform such a monumental task on a reagent that is FDA licensed. You can validate or qualify a piece of equipment and validate a process or method, but I do not believe using the term "validate" with a blood grouping reagent is appropriate. Manufacturers are required as part of licensing to perform an enormous amount of testing (clinical trials) to enusre the reagent reacts as intended. The testing performed by any user would be minimal compared to what the manufacturer has already done. It would be a waste of valuable reagent and not add any value to test results. If you follow the manufacturers test instructions properly and expected results are obtained each day with known postive and negative controls, test results are valid. To meet Quality System requirements you should have a process to evaluate/qualify suppliers.
-
It appears everyone is on the same path. Of course the CFR does not specifically tell you what should be in a supplier qualification program, but that you must have one. The basic principles in our procedure includes: 1. Assess the supplier to determine if critical or non-critical. If non-critical document rationale. If critical evaluate as below. 2. Send the supplier a questionaire with basic questions about their operation including ISO certification 3. Evaluate the questionaire and search for FDA warning letters, need for a site audit, etc 4. Track the suppliers performance by documenting any problems and resolution in the suppliers file 5. Periodically review suppliers file for performance issues to determine continued approval This has worked for us so far. By tracking supplier performance we have needed to discontinue using a couple of suppliers that continued to give us problems. Mostly because their corrective actions were insufficient.
-
If both controls were run on the same day I would also question the results obtained before the second control failed. There would need to be more investigation as to why the controls failed to be able to evaluate previous test results. One could argue the need to run controls at the begining and the end of a run when performing automated testing. As long as both sets of controls are acceptable there should be no question to the accuracy of the results. I have seen this done for infectious disease testing.
-
I have never heard of bacterial contamination causing spontaneous agglutination or something like polyagglutination, but I guess anything is possible depending on the type of microorganism. From a quality point of view, the validity of any test results obtained after a control fails should be questioned.
-
I would suspect improper handling of reagents if bacteria contaimination was actually proven. In the US the FDA requires that licensed blood grouping reagents undergo preservative effectiveness testing. This is usually done according to USP standards where the product is challenged with 1X10 6 concentration of microorganisms. A colony count is then performed over a 28 day period to determine log reduction of viable organisms. The preservative helps control organism growth over the shelf life of the product since these reagents would be considered a multi-dose device. I think it would take a considerable amount of bacterial contamination to make an Anti-D reagent stop working since they are usually very potent reagents. For manual typing reagents I go with running controls daily. On automated instruments I would rely on the instrument manufacturer's recommendation for control frequency.
-
Reagent manufacturers must have data to fully support any claims in their package insert. They should have stability data over the shelf-life of the product. If the manufacturer really covered all the bases they should also have data on some failure mode testing. For example; using the same reagent container over multiple days to show that it works, keeping the reagent at room temperature for extended time and even shipping validation to prove the product can handle temperature excursions during shipment. Contacting them to see if they have additional data may not be a bad idea. They cannot put everything they have done in the package insert or it would be a book. Most reagents have a preservative in them to help prevent microbial contaimination. For blood grouping reagents that is traditionally sodium azide. This preservative is very effective in microbial control. We also need to rely on the controls that are run. They are run for a good reason. Expected results obtained with proper postive and negative controls verify reagents are performing as expected.
-
Performing daily QC of screening cells using a weak antibody is a good idea and necessary. But, QC of panel cells. This would really seem like a huge task. I would not know were to begin. What antigens do you type for on the panel cells. Are some more liable than others that could indicate deterioration or would you need to test them all. This would use a significant amount of antisera and take a lot of valuable time. I would suggest panel cells do not need to be QC'd. They are used to determine antibody specificity. If you use multiple cells for rule outs, it would be a low risk. Other things like confirming the patient's red cells are negative for the antibody, typing units for the antigen and the actual crossmatch help prevent problems. Of course using expiring panel cells has been a bit of a controversy for years.
-
Evaluation of Anti-A, -A,B, -B, and -D Reagents
JHH1999 replied to ckcheng's topic in Education / Quality
Manufacturers of blood grouping reagents including Anti-A, -B, -AB and -D are required to meet a potency standard and must be specific. Some have reference standards and others have minimal potency requirements listed in 21 CFR 660.22. Most manufacturers are required to submit samples of product to the FDA for release. This link should take you to the FDA website and the document describing the methods recommended by FDA to evaluate performance of these reagents. http://www.fda.gov/cber/gdlns/bldreagent.pdf The testing manufacturers perform is very extensive and there would be no reason for end users to repeat this testing to qualify FDA licensed reagents. Performing daily QC using manufacturer recommended controls should be sufficient. -
I would suggest it would be good practice (GMP) to include the lot number and the expiration date of the saline on any container it is transferred to for use. Internal and external auditors would expect to see traceability of this information to a scondary conatiner. If you have not been cited yet, it could be an issue in the future since more Quality System Regulations are being applied to blood banks.