Jump to content

D-dimer QC issues - IL


SMILLER

Recommended Posts

We have been having problems for the last month or so with D-dimer QC on our TOPS analyzers.  We are using the regular HemosiL reagents and controls, and no matter what we do, the QC runs with a Mean close to the upper insert range limit on both controls.  Does anyone else have this problem?  

We have been using these analyzers and IL products for over a year and now this happens. We are wondering if the formulation of the QC has changed.  Once reconstituted, they are supposed to be good for a month in the fridge -- it appears to us that they may be only reliable for about a day. 

Thanks, Scott

Link to comment
Share on other sites

When I started with the DDHS500 line a few years back I had a horrible time with controls. That 30 day claim is junk - and I thought they were going to work on changing that wording in the package insert. I could only get 3 days out of mine. They did seem to work out the issue I was having, though.  I believe it was a combination reagent stability AND qc stability.  (We ended up switching to BioRad controls simply because in NYS I have to have a "negative" control, and IL doesn't make one that low yet.)  But in all the years of using IL, that is the ONLY problem I've had. The TOP analyzers are rock stars! (and I just got my new 550's in!)

The best thing to do is document EVERYTHING. I had a log at the instrument where I had the techs write down every time they changed reagent or made up new qc and why (empty vs. qc trouble). And then I would print the Levy-Jennings and send them in.

Do you submit monthly to the AccuTrak? If everyone else is running high, that is in your favor.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.