For those who used or are currently using the Vision, did you extent the reagent red cell on-board expiration past the recommended 5 day stability? If so, can you share some tips for validation.

 

Thank you

 

Bman

  • 1 month later...
  • Replies 5
  • Views 2.2k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • I believe the antibiotics and antifungals start to degrade after 5 days onboard rather rapidly.  The excessive ? reactions are not worth the hassle.  

comment_87548

We do not extend the expiration date but love to know if others find you can because we never use the entire bottle and it seems like a big waste of money.

comment_87562

we use ours too quickly - but we would not extend......they seem to have more ?? the longer they are on the machine - the more they are used.

comment_87585

I believe the antibiotics and antifungals start to degrade after 5 days onboard rather rapidly.  The excessive ? reactions are not worth the hassle.  

comment_87603

We don't use them up after 5 days but with the low humidity here the suspension is quite thick after 5 days even with the evap caps.  You can tell a difference in the cell button size from day 1 to day 5 but it must be minimal enough that the system doesn't flag it.  If someone forgets to change them, you're guaranteed to get TMC errors "too much cells."

  • 6 months later...
comment_89656

i know this is late, but we are a small volume lab and we validated onboard stability up to 10 days (though we routinely change them every 7 days for ease of scheduling).

first thing, we are also in a low humidity area - the evap caps must be on there in such a way that the flaps/flanges are fully closed - too tight and they sort of stand up and open - a few days like that and you will definitely get TMC (too many cells) errors due to concentration.

also, note the verbiage used regarding the onboard stability of the cells - it doesn't say that it's 5 days, it says that it hasn't been validated beyond 5 days.

we simply ran QC side by side each day for 10 days with fresher cells. all went well - no QC issues, no "degradation" or visible issues. medical director approved the validation. done deal.