Centrifuge timer checks


Byfaith

Recommended Posts

Our BioMed department checks our RPMs and timers quarterly, as required by NYS.  Should we be asking them to verify timers for every increment we use?  (ie 20sec AHG,  25sec IS, 45sec Wash)  Also, we found out they are just using their phones to check timer, not a validated stopwatch.

Thoughts?

Link to comment
Share on other sites

  • 4 months later...

As for how many points to calibrate, my personal opinion is to do three points.  The lowest should be below the lowest range you'll use, the highest above, and then a midpoint.  This is not perfect, but it helps prove linearity.  Many places only do one or two points, though.  It also depends on the significance of what you're measuring.  For instance, a 20-second AHG spin... does it really matter if it's 20.00001 sec or 20.00002?  Of course not.  That's what leads some people only to do one or two points.

When you say they're using their phone, do you mean a timer inside the phone?  I stick by my original statement above; it's irrelevant for what we do as med techs... but..., in this case, as your assessor / inspector / regulator, I will want to see that you are using a traceable device for these measurements.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.