Posted March 7, 20196 yr comment_76185 I have two laboratories that each has a Sedimat 15 for performing automated erythrocyte sedimentation rates. Does anyone know what the target percentage should be for correlation studies between the two analyzers?
August 21, 20195 yr comment_78053 Were you able to get good correlation? I seem to have trouble at the high end. When the result is around 90 or greater, I'm not getting good correlation.
August 21, 20195 yr Author comment_78056 Well, not at first. When I started my correlation studies I was sending blind samples to my techs at the 2 locations where my Sedimat 15 analyzers are located. Some results matched perfectly while some were way off. So I took 5 samples and set them up at both locations myself. And guess what? They matched perfectly. I then found out that my lack of correlation was more of an operator issue rather than the instruments not matching. I think the biggest factor was that some techs were not pressing the pipette all the way down into the pink vial. With all that being said my sample with the highest result during my last study yielded results of 65 and 70. My criteria for acceptance is as follows: Result Target Value 1-15 +/- 4 16-30 +/- 6 30-50 +/- 8 >50 +/- 10 I hope this helps!
September 9, 20195 yr comment_78295 May I ask where you found your criteria? I'm sure my supervisor will want that information.
September 10, 20195 yr Author comment_78307 I loosely used the MLE Proficiency Test standard of +/- 2 standard deviations. Since I only have 2 analyzers I came up with ranges based on the values of the results.
September 10, 20195 yr comment_78308 Thanks Chris! After I sent my question to you, I noticed someone down below said they used CAP's standard for proficiency testing of +/- 3 SD. Hopefully, I'll get better results. I went as far as making sure the counter top was level lol!