  1. Are your panels tested in tube or solid phase? 

     

    If in tube, there's a possibility that this antibody may have formed against a constituent in your reagent cells

    based on the fact that all other allos are ruled out and you are seeing are seeing strong reactions with no pattern.

     

    I have seen these antibodies titer high.

     

    Try testing panels with washed reagent cells to see if it gets rid of the reactivity.

     

  2. There is an excellent article in Immunohematology, Vol 16, #4,2000 that decribes

    the validation of this procedure by the ARC in Columbus, OH.  It compard 26 samples

    using IgG blocking, glycine acid EDTA, and chloroquine diphosphate treated red

    blood cells in typing for s,S,Fya,Fyb,Jka, and Jkb.  The procedure works well with

    samples that have a w+ to 2+ DAT.  One sample tested Fyb+ with the blocking method

    and Fyb- with chloroquine and glycine acid EDTA treated cells.  I remember reading in

    the insert that chloroquine diphosphate treatment of red cells can weaken the Fyb antigen.

     

     

     

  3. Does anyone have any procedures that they can share for making up student specimens, either spiking plasma samples with antibodies but even more so, making up specimens for doing eluates?  If there are some on this site already, sorry for the repeat, but I can't find them.

    To demonstrate mixed field in forward typing, mix together the contents of one segment

    from a grp A blood unit and one segment from a grp O blood unit.

  4. Does anyone transfuse plasma products based only on patient ABO/Rh history? For example, if they had T & S done 6 months ago and they were a B+. Today they come in and the physician orders FFP to be transfused. Would you require a new ABO/Rh or would you issue plasma based on the B+ from 6 months ago?

    We require a new sample for each new admission. In this case a sample would

    be drawn for a ABO/Rh type.

  5. We wash a suspension of cord red blood cells maunually 3x.

    From these washed cells we do an ABO and Rh type in tube.

    Also, from the washed cell suspension, 1 and 2 drops of it are

    placed into 2 respective tubes, wash 3-4 times in the cells washer

    and then add 2 drops of Anti-IgG to eaxh tube for the DAT.

  6. We do a type on screen on neonates for possible transfusion. We transfuse type specific blood so we also

    carry the Baby's plasma to the AHG phase when mom is grp O and baby is grp A, B, or A,B to check for presence of isohemagglutinin.

    If present we give grp O. The type and screen sample is good for the lenght of stay the infant is in house which can go up

    to 4 months as an expiration.

    A DAT must me performed if a previous cord sample was never done. If the screen is positive due to an alloantibody we try and identify it in the neonate's sample. If QNS, we use mom's plasma for ID and x-match.

  7. OES I believe stands for Orhto Enhancenent Solution which is a low ionic strength solution. Often times warm autoantibodies

    are seen to be reactive using gel, peg, and'or solid phase but do not react in a low ionic strength medium. By testing patient

    plasma with it one can use it to rule out any coexisting allosntibodies. Another way to test patient plasma is by using 2 drops

    of palsma with one drop od reagent red blood cell. Incubate 30-60' at 37C, wash 4X, add Anti-IgG. warm autoantibodies often

    do not react using this technique.

  8. The reference lab I worked in performed all initial work ups using LISS. If negative, and a cold was suspected then a ficin panel

    was tested. Also if LISS panel was negative a PEG screen was tested.

    For a short while prior to this, only PEG was used and a 4:1 screen at IS, RT and 4C were set up as the initial testing.

    A few of the submitting hospitals were requesting that our reference lab should acquire the gel methodology so we could duplicate

    their results. Gel/ProVue were subsequently acquired in the ref lab. Our experience also in this reference lab was that the requesting facility would detect nnspecific reactions in gel and solid phase and we would get negative results using LISS, PEG, and

    ficin treated reagent red cells.

    My experience now with solid phase has been seeing what looks like an antibody detected reacting 3-4+ with all 14 wells with negative and then negative reactions using gel and tube. When rouleaux has been rule out, I suspect that there is an

    antibody in the patient sample that is reacting against some unknown entitiy(stroma or chemical) that is affixed to the bottom

    of the u-shaped well. Also with solid phase, newly forming antibodies whose make up is largely composed of the IgM isotype

    (primary immune response) are on a few occasions missed using solid phase and then shown to demonstrate a clear pattern of reactivity in tube/LISS in which there is that valuable 37C reading. Also, the newly forming antibodies to not have the proper affinity

    in the hypervariable reagions of the FAB portion of the immunoglobulin which does not result in its proper fit with its epitope. This

    latter characteristic would result in a hard to or too weak to identify pattern using any methodology.

    Bottom line is to use all methodolgy that is available for ID purposes. If time and cost start to be an issue then flow charting

    based on what type of reactions seen with what methodolgy is of primary importance.

  9. The reference lab I worked in performed all initial work ups using LISS. If negative, and a cold was suspected then a ficin panel

    was tested. Also if LISS panel was negative a PEG screen was tested.

    For a short while prior to this, only PEG was used and a 4:1 screen at IS, RT and 4C were set up as the initial testing.

    A few of the submitting hospitals were requesting that our reference lab should acquire the gel methodology so we could duplicate

    their results. Gel/ProVue were subsequently acquired in the ref lab. Our experience also in this reference lab was that the requesting facility would detect nnspecific reactions in gel and solid phase and we would get negative results using LISS, PEG, and

    ficin treated reagent red cells.

    My experience now with solid phase has been seeing what looks like an antibody detected reacting 3-4+ with all 14 wells with negative and then negative reactions using gel and tube. When rouleaux has been rule out, I suspect that there is an

    antibody in the patient sample that is reacting against some unknown entitiy(stroma or chemical) that is affixed to the bottom

    of the u-shaped well. Also with solid phase, newly forming antibodies whose make up is largely composed of the IgM isotype

    (primary immune response) are on a few occasions missed using solid phase and then shown to demonstrate a clear pattern of reactivity in tube/LISS in which there is that valuable 37C reading. Also, the newly forming antibodies to not have the proper affinity

    in the hypervariable reagions of the FAB portion of the immunoglobulin which does not result in its proper fit with its epitope. This

    latter characteristic would result in a hard to or too weak to identify pattern using any methodology.

    Bottom line is to use all methodolgy that is available for ID purposes. If time and cost start to be an issue then flow charting

    based on what type of reactions seen with what methodolgy is of primary importance.

  10. Hi Shelly,

    We do the same, however, we recently changed the last step in that we no longer need a physicians signature.

    Instead of the signature, we enter a charted comment in the patients profile that allows the blood banker to call

    this warm autoaby as a critical result to the RN. Any units that are least incompatible due to a warm auto or HTLA

    are considered critical results.

×
×