Dear Mini-Me et al: The primary purpose of the indicator is to provide some form of visual assurance, right on the blood bag, that the blood product has been irradiated. The secondary purpose is to provide a qualitative indication of the degree of radiation exposure. Some indicators provide a semi-quantitative indication of the level of dose: a color matching scheme is used whereby you can visually ascertain whether the absorbed dose lies between, say, 15 Gy and 50 Gy (i.e., RadTag brand indicators). However, indicators are not dosimeters and cannot be used as a substitute for quantitative dosimetrty, nor do they elimninate the need for other process-control mesures such as controled segregation of nonirradiated from irradiated products. When selecting or validating indicators for blood product irradiation, you need to consider several factors. Chief among these are: 1 - suitability and convenience of use: ability to remain attached and withstand the stresses of irradiation as well as their integrity/stabilty w.r.t. the environmental conditions they are subjected to (i.e., extremes in temperature, relative humidity, exposure to ozone, UV, etc.). 2 - response: must be appropriate for the range of dose, dose rate, radiation energy, and environmental conditions experienced by the irradiated product. Once you have selected an indicator (one designed specifically for x-ray based units or one specific to gamma based units), and can ensure that you are only using them under the manufacturer's stated conditions of use, then your validation protocol should include the verification of the indicator's RESPONSE for your particular type of indicator as verified in YOUR irradiator under YOUR environmental conditions. This also applies when changing lots within a given brand or type. The response must be such that it does not indicate that an absorbed dose level of 15 Gy has been achieved when, say, only 10 Gy has been delivered. That is why I proposed the irradiation protocol and dose levels that I did. In order to test the response at accurate dose levels, you will most likely have to ensure that the sample chamber is completely filled with water-equivalent material and that you place the indicators in the centre. These are the conditions under which your dose rate has been measured and for which your timer setting has been set/calibrated to. I agree with your comment that indicators should be treated as reagents - they are. Indicators typically have a limited shelf life and must be properly stored (e.g., refrigerated, kept out of UV/sunlight). Given enough time, most indicators will age to the point where their integrity is compromised and so it is important to respect the manufacturer's expiration labels. Best regards, Grant Nixon, Ph.D., P.Phys.