Thanks ABBWalker, as you say this is very much open to interpretation, and as we discussed earlier (privately), I do have concerns about the 200-250ml volumes used for core readings. Whatever validations/ testing is being performed should always be based on looking at the 'worst case' scenario, in this case if paediatric blood units are being stored, then the core readings should be performed in 50ml liquid. If a paediatric pack is constantly exposed to air temps above 6'C, then it is possible that the actual component itself will reflect this temperature and not the required one core one. How do we know to what level of harm this may cause a neonate? If small vial products such as anti-D Ig is being stored (?3ml vol), how long would the internal core temperature of this volume remain at 2-8'C (which is different to red cells in uk= 2-6'C), if the air temp of the unit fluctuates regularly. Does air temperature of a cabinet= the core temperature of small volume products?.....this is something that may need to be addressed by the mapping. I am very surprised that pharmaceuticals/ hospital pharmacy dept etc have not addressed this. Some medications can be severly affected by temperature variations such that they may not be of optimum benefit to a patient if stored incorrectly........about time hospitals mapped all their medicines fridges????