Everything posted by shelleyk482
-
-80C thermometers
I ordered an ultra-low thermometer for our plasma freezer that should have been in 30 ml of propylene glycol but when we got it, it was in sand instead. Thinking that the wrong thermometer was put in the box (we verified manufacturer's numbers) we called Cardinal and were told that these now only come in the sand. My question: would this be an appropriate thermometer for Blood Bank freezers? Are the heat transmission properties considered to be the same for sand as it is for frozen products?
-
Light box for Manual Capture system
I ended up finding a battery-powered light box for $42.00. It still probably doesn't have the UL certification (it's not mentioned in the description) but I convinced Biomed that it posed little to no danger (lower voltage) and since the manual method is a backup to a backup and hopefully will never really be used (Tango is our primary method, Echo will be our back-up and if both of these fail at the same time then we will use the manual capture) I was able to convince them to approve it. I was very frustrated but as a Blood Banker, I know a lot of nurses that think I over-interpret regulations so I guess turn-about is fair play.
-
Light box for Manual Capture system
We just received the ImmucorGamma manual capture system and our BioMed department will not approve the light box viewer for use. They are concerned that the "approved by UIL" label is on the electrical element inside the box and not on the outside so their interpretation is that the entire viewer is not UIL approved as required by the Life-Safety code. There is also just a 2-pronged plug that doesn't have a ground and the unit is not double insulated. Has anyone else had this problem? Does anyone know where to get a hospital-safe view box? I did a Google search and all I came up with were expensive Rh view boxes with heaters in them and most places had discontinued these anyway.
-
XM for a newborn with mom's specimen
I would just give antigen negative units. Baby can't make his/her own antibody so if you've identified the anti-E in Mom, ruled out other antibodies and have proven that baby does have the passive Anti-E, my understanding is that you just have to give antigen negative units until the passive antibody is no longer demonstrable in baby's system, no need for any crossmatch. For an anti-E, I would probably just give E-negative units without/XM until the baby is 4 months old or no longer needs blood (which ever happens first) without repeating any tests to see if the Anti-E is still demonstrable; it doesn't hurt to give E-negative units even if the antibody is out of the baby's system.
-
ALL automation???
The more I've thought about it, the more I like the idea of the 2 vendors. This would give me a second automated methodology rather than an automated & a manual. It is highly unlikely that any patient specimen would not be able to be analyzed by both methodology so if I get "iffy" reactions using the primary then I could put it on the secondary. If I have problems with both methodologies (either 2 automated or 1 automated and 1 manual), the specimen would be sent out to the blood center. As far as the generalists using the 2 different analyzers, they already use several different machines in the other areas of the lab, so this really doesn't seem to be a concern for them. We are already maintaining competencies on 3 "platforms": automation, gel & tube. This would just take the gel (and probably the tube) out of the mix. I do appreciate the input. This is allowing me to "think out loud".
-
ALL automation???
We currently have one blood bank analyzer and use manual gel for backup. I have been offered the opportunity to bring a 2nd piece of automation into the blood bank- a different manufacturer. At first, I totally dismissed the idea but now I'm actually trying to look at the pro's and con's and I would really like some other perspectives on this. My idea at the moment would be to use the analyzer we have now as the primary and to use the 2nd for retypes, as backup for the primary and as secondary methodology for complex samples (we don't do elutions or absorptions). If we have the two analyzers, I don't think I would need to keep a manual backup system. This idea is both intriguing (from a manager's aspect) and disturbing (as a blood banker). All of our techs are generalists and the ones I've discussed this idea with, don't seem to have a problem with the idea of doing maintenance on another machine; they seem to think that this would be faster than doing the QC on the manual gel system. I appreciate all input, ideas, etc.
-
alarm testing for refrigerators/freezers
Last time I checked with CAP (and it's been a couple of years) they still wanted the actual physical check and not the electronic check. If you are CAP accrediated, I would definitely check with them before I changed my requirements.
-
Help ! Issuing soon to expire blood components.
The Circular of Information states "Transfusion should be started before component expiration and completed within 4 hours". (pg 5, #12, Dec 2009). We would require Medical Director approval to start any transfusion after the expiration time based on the Circular statement. I hate to waste product but I would hate more to have to justify to an inspector or court why I didn't follow this for a "readily" available product.
-
Type and Screen and Hold-Do labs use this order?
We use it for our outpatient cancer center & our L&D. The cancer center draws a "Hold" sample at the same time they draw the CBC on their outpatient; then if they determine they need to transfuse, we don't have to restick the patient. Very few of our L&D patients need blood bank testing but by drawing the "hold", we assure that we don't have to try to draw during a traumatic birth. We no longer let OR use this order; we apply a blood bank ID band when we draw a "hold" sample and we had problems with OR assuming that if there was a blood bank ID band on the patient that the patient had RBC's available.
-
Giving group AB platelets to non group AB neonate
Check out page 10 of "Pediatric Transfusion: A Physician's Handbook", 2nd edition. It is a table that gives this information. The column does not say "platelets" specifically, it says "Plasma-Containing Components" which does include platelets.
-
Verbal Orders for Blood/Blood Products
We accept verbal orders during emergency situations- trauma or OR. This generally happens with uncrossmatched for trauma or OR patients who start massively bleeding and we run out of what was originally set up. We try to get a computerized order but will not delay getting product to the patient if we can't get immediate compliance from nursing.
-
Helmer freezer alarm checks
I sent this question to CAP a couple of years ago and was told that this was just an electronic check and did not meet the requirements. I have not heard that their position has changed so we continue to do the maual method of alarm checks.
-
Cerner Millenium and Bridge
We are a Cerner Millenium system and I just heard about the Bridge product that can be used on Cerner for tranfusion related documentation. Does anyone have experience with this program; if so, how do you feel about it? Is it a good system, what problems have you seen, etc?
-
QNS Specimens for Blood Bank
Our policy is that a sample collected at a different time requires all pretransfusion testing if it is going to be used for crossmatching. We do not automatically reject a short sample; we try to look at all of the factors that L106 mentioned above before asking for a recollect. If there is any reasonable indication that we may need need sample than we get, then I prefer that my staff get a recollection rather than run the risk of running out of sample during a dynamic situation.
-
IS with AHG crossmatch
In which section of the CLIA regulations did you find this requirement? I've scanned through document and evidently missed it. Thanks.
-
IS with AHG crossmatch
The 16th edition of the Technical Manual states (pg 453) "Although tube and column agglutination AHG crossmatches will also detect ABO incompatibility, the AHG crossmatch in solid phase may not. An additional procedure is required to verify ABO compatibility, such as the IS or computer crossmatch". So my understanding of this is that the only time you need to do the IS in addition to the AHG phase is if you are using solid phase technology to do your crossmatch.
-
FDA reportable?
If this was not a planned deviation from SOP with prior approval by the Medical Director, then my opinion is that this is an FDA reportable.
-
Filter size for neonates
We also use a sterile connecting device to attach the syringe to the parent bag. We only do a few aliquots per month (3-5) but I think the SCD is worth investing in. The BB staff and the nursing staff all feel that we are using a safer product since we went to using the SCD. The initial cost is fairly significant but I was able to get it through based on patient safety; plus the SCD lasts forever and rarely (if ever) needs repair.
-
minimum volume of FFP?
Blood Transfusion Therapy: A Physician's Handbook (9th ed, 2008) on page 31 states "When plasma is given for coagulation factor replacement, the dose is 10 to 20 mL/kg (3 to 6 units in an adult). This dose should be expected to increase the level of coagulation factors by 20% immediately after infusion". I hate to see us go this direction because I'm afraid that wastage will go up if physicians start asking for an exact volume. Adult products have routinely been given by unit and not by exact dosing amounts and in over 30 years of doing this, I have yet to see a complication that I could reasonably attribute to this practice. Those of you who give products to adults based on dosage (mL/kg) criteria, what did this practice do to your wastage rate? For those of you considering going to this practice on adults, what is the reasoning for considering the change?
-
Filter size for neonates
In the Children's Hospital I worked at a couple of years ago, our neonatologists insisted on a 40 micron filter. Their claim was that microclots could be detrimental to the vasculature of their small patients. I was not able to find any literature to support this and they never furnished any but it's what they wanted so it was what we used. Pall used to have a list on their website (they may still have it) of the facilities that used their syringe aliquot system and when I called some of those facilities, I found that most used the standard 170 micron filter. If the list is still available, you may want to do a poll of these facilities and see what the current standard practice is.
-
Off site clinic transfusions
If you choose to provide compatibility testing and products to a non-affiliated facility, make sure that the contract is clearly written as to which facility is liable for which functions. You will want to clearly delineate that your responsibility begins and ends with the testing performed at your facility (i.e. the specimen you received is compatible with the product you issued); if employees from your facility do not collect the pretransfusion specimen, you cannot claim that the product furnished is compatible with the patient. Responsibility for proper patient identification has to be expressly assigned to the facility collecting the specimen & transfusing the product. I have also insisted on doing periodic inservices with the contract facility concerning proper patient identification, specimen collection, handling of blood products, etc. You will need a policy/procedure that clearly defines labelling and specimen requirements and you must insist that it be followed. You will also probably want to define how they transport the product (type of cooler needed, etc); I've always insisted that they bring an appropriate transport container and do not issue if the transport container is not brought to the lab at the time of pick-up. The policy should also state your return policy if a product is not used for some reason; I have never allowed return of a product that went outside of our control, if you do, get with your blood center because this may affect your ability to return consigned products. You will also need to decide how you will handle possible transfusion reactions. I have worked in several different places that have these types of contracts and it hasn't been a problem. When setting up the process, you need to make sure that you think out every single step and every possible liability and clearly delineate this in the contract &/or policy/procedure.
-
minimum volume of FFP?
I don't know that there is a minimum- never run across one. We issue by # of units for adults. Only do by mls for infants or small children; the clinician determines volume to be given.
-
Thawed plasma outdate
It is easier for inventory management if all thawed plasma expires at the same time. When the expiration times stagger, the techs have to be more aware of when a particular product may expire. We have recently had 2 instances where a product expired during a shift and less than 30 minutes later we had an order and could have used the product if we were using the 2359 expiration. With a 2359 expiration, it is less likely that this type of scenarion will happen during the night.
-
Thawed plasma outdate
You can use either. For products with an expiration of less than or equal to 72 hours, you must use the exact time. For products with an expiration of more than 72 hours, you can use the number of days @2359 (Technical Manual, 16th ed, pg 219). We have been using the 120 hrs but are in the process of changing to the 5 days @2359.
-
Emergency Transfusion for Neonates
We have a small NICU and we do not use CMVneg or irradiated for our neonates- we use the freshest AS-3 unit on our shelf. Our sickest babies get transferred out but while they are here, they get CMV safe (leukocyte reducted) products not CMV tested. Irradiation Guidelines in the Technical Manual states that irradiation should be done for premature infants weighing less than 1200 grams at birth so for preemies larger than that, irradiation is not required.