    Red Blood Cells reacted to webersl in Settings of centrifuge to separate blood samples   
    I agree with all the posters here who ask you to return to SOP until you have convinced your superiors and they have changed the SOP. Here is how I switched from a 10 minute spin to a 2 minute spin (I am a supervisor):
    For each of 20 specimens (10 w a pos antibody screen, 10 with a negative antibody screen) I spun according to the SOP (10 min) and on a worksheet I commented as to the visual appearance, including clarity, hemolysis, and the sharpness of the interface between cells and plasma. Then I tested the specimens according to SOP and reported. Then, I completely resuspended the specimens on a tube rocker and respun them at the new setting (StatSpin3) for 2 min. After the spin I again recorded the visual apearance and then repeated all the testing and recorded the results on a worksheet (did not report). I then compared the apearance and results. There was no difference in any of the visuals or results, so I considered the new process to be valid. I wrote up the mini-study and sent it to my Medical Director for approval to change the SOP accordingly. Once approved, I began making the SOP changes. Then, staff read the new SOP, were trained on the new centrifuge settings, demonstrated competency. Then I announced to all staff that we would 'go live' with the new settings on a certain date. I realize your lab did not perform a comparison of any kind when they got their new centrifuge, but they should have. If you cannot find a way to convince them, I am sorry for you, but you MUST continue to follow SOP as written, or speak with your Medical Director (physician in charge of the lab) about your scientific concerns so that a proper change may occur. Best of luck.
  2. Like
    Red Blood Cells reacted to tbostock in Settings of centrifuge to separate blood samples   
    We use Stat Spin centrifuges that spin in 3 minutes. We validated it and made sure there were no problems in gel testing. Also, Stat Spin will provide you with a validation study that was published about the 3 minute spin time at higher RPMs for Blood Bank specimens.
  3. Like
  4. Like
    Red Blood Cells reacted to Eagle Eye in Secondary Bacterial Testing for Platelets   
    PAS ---we do not need to do testing ...but can not be extended to 7 days..
    I thought there was a limitation that supplier could not collect 3 bags?
     
×
×