We validate all of our coolers empty, with two (mock) units and full to capacity.

Most of our coolers leave with 2 or 4 units, but sometimes we do an exchange and issue 2 or 3 full coolers.

I suggest you validate your coolers to mimic how they will be used.

Link to comment
Share on other sites

 I validate the coolers with the minimum and maximum volume the cooler can hold. In your case that would be 1 unit and 8 unit. My reason is to ensure the 1 unit doesn't freezer and the 8 units are not to warm.

In the patient care area, there will be times only 1 units is left because they are using them. And you may send the cooler out with eight. The goal is to qualify the equipment at low and high ends.