Washed / deglyced RBC QC


Cliff

Recommended Posts

Same colorimeter we use. I bet they sold/gave away thousands of those things. Our platelet apheresis nurses use one on all our SDP components. We also have an analyzer in our chemistry department (I still need my GPS to find the place) that measures plasma free hemoglobin.

BC

Link to comment
Share on other sites

I've always thought this was something we made up to make the inspectors happier.........but here you go. Hct of both pre and post deglyced unit...calculate % RBC recovery.......also "simulated transfusion" (don't laugh) of .5ml post sample mixed with 10ml of .9% saline. use comparator to determine "hgb".....and of course we use it during every wash/deglyc. procedure.

Link to comment
Share on other sites

Bob,

We have two 2991s -- one that is the old "peg board" style, but we also have a new one where the "pegs" have been replaced by a digital display (the knobs are still there). We bought the new one just a few years ago, and everyone who was trained on the old one still prefers that.

We perform the following tests on the supernatant prior to release: Washed RBCs: supernatant for protein using a urinalysis dipstick; Deglyced RBCs: we also use the Haemonetics Color Comparator. We used to test the Refractive Index or Specific Gravity on the supernatant, but since our refractometer died we have had our reference lab do the "simulated" transfusion described in the Technical Manual.

We also perform %RBC recovery for each procedure quarterly.

Heather

Link to comment
Share on other sites

  • 4 months later...

To all of you who do dipsticks for protein. Do you have an SOP for that and do you run controls? Just wondering as I can anticipate the next question from the assessor. Did any of you "validate" the process to show that it consistently resulted in a protein below detectable levels? If so, did that not fly with the assessor?

Link to comment
Share on other sites

  • 1 month later...

Sshel55: Yes, we have an SOP and we run a positive and negative control on each day of use. We validated the use of the dipsticks. We did not use the dipstick when we validated the instruments because we weren't doing protein QC when the instruments were implemented.

P.S. We just implemented an Atago digital refractometer (the same one seen on the last page of the MarketLab catalog for Spring 2008 even though that's not where we bought it from). It is DELIGHTFUL and I would highly recommend it to anyone!!

Link to comment
Share on other sites

Thanks for the reply, Heather. What is the protein content that you all consider acceptable in the final wash supernatant and how did you determine the lacceptable level?

I'll check out the refractometer. It might come in handy for protein levels for serial plasmapheresis donors.

Sherry Sheldon

Link to comment
Share on other sites

We say that the "Negative" and "trace" values on the dipsticks are acceptable. I don't remember exactly what our rationale was, but I think those were the values that Production and QA thought that would qualify as "significantly reduced" or "minimal" residual protein.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.