Establishing reference ranges for Factor VIII testing


kholshoe

Recommended Posts

Hi everyone! 

 

We currently perform Factor VIII testing, but we do a very small volume.  The testing isn't batched, but we just do them as they are ordered.  Looking back over the last few months - we normally perform <5/month.  (!!)

 

With new lots of Factor VIII deficient plasma - what's the best way to establish new reference ranges for this test?  Any thoughts?

Link to comment
Share on other sites

The standard way is to calibrate you machine based on already established control material, run your new batch of control at least 10 times establishing the values are in the acceptable range provided and then set your mean and standard deviations from this.

Be very careful about using the standard deviations provided by the manufacturers as some of these can actually be 3sd not 2! I was horrified when I found out...

Link to comment
Share on other sites

your original question mentioned "reference ranges" - just to clarify, you don't need to validate reference ranges for patient results when a lot number switches.  regulatory agencies prefer you create your own control ranges, but there is also an "out" for tests that are not performed frequently enough (or for tests in which it would be very expensive) to establish these ranges.

for any of my special coag testing, we use assayed controls, we use the manufacturer mean, but we do tighten the acceptable range to 2sd.  as auntie-d mentioned, many published ranges are, indeed, 3sd - and (are you sitting down?) IL published ranges are a whopping FOUR sd!

Link to comment
Share on other sites

Thanks everyone!  We have a procedure in place to establish control means and ranges, but my question was actually referring to the test reference range itself.  Ie...how do you establish a new "normal" reportable range with a new lot of reagent?  I'm not sure I'm being clear =P

 

Nziegler - when you stated: " you don't need to validate reference ranges for patient results when a lot number switches" - are you referring to control ranges or assay ranges?  Can you point me to some documenation supporting this?  That would be great!

 

Thanks again =)

Link to comment
Share on other sites

We re-validate reference ranges when we switch lot numbers. (But we do not run factors here.) We run at least 25 presumably normal (hospital associate volunteers not on birth control, coumadin, etc.) to establish a reference range. In general, if the range is very close to what we have been using with the old lot number, our pathologist will not chamge it.

Scott

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.