Anyone want to help a newbie supervisor?


kholshoe

Recommended Posts

Hello all!

 

My name is Kristen and I'm the supervisor of the hematology/coagulation department at our laboratory.  I have about 9 years experience as a generalist on several shifts, but I am brand-spankin' new to this role.

 

We have had issues with our sed rate analyzer for a few weeks.  The company is sending us a new one tomorrow (same model, etc...). 

 

I realize things like accuracy, precision, and analytical range must be established.  However, is it any different for this analyzer since it is the same method, model, etc..?  And, how would your facility go about establishing these things in a situation like this?

 

I was planning on running QC on both the old and new analyzer for at least one week - this should prove accuracy sufficiently.  I was also going to run a batch of patient samples as a comparison on the old vs the new analyzer...

 

How do you prove analytical range?  Would you try to establish reproducibility as well?

 

I'd appreciate any help on this.  My resources are, unfortunately, vague.

Link to comment
Share on other sites

For reproducability, I would just run several ESRs on the same set of specimens.  For reportable range?  I guess you could just say it's whatever is readable on the analyzer.

 

What's wierd, I think, about ESRs is that they are basically a manual method with an automated reader attatched, and all of this stuff seems kindof like overkill.

 

Scott

Link to comment
Share on other sites

I'm thinking a reproducibility study on this might just be overkill as well.  Several of the other supervisors in the lab suggested running QC on both the old and the new analyzer for one week.  And, also running some patient comparisons (but they felt that 10 samples should be sufficient). 

 

Nothing is changing in my procedure/policy because the instrument is exactly the same model, etc.  I guess I'd feel like more studies were necessary if we were moving from one method of analysis to an entirely new method of analysis.

 

I agree that alot of this seems silly....a sed rate analyzer is truly just a "reader" like you said.  Thanks for your opinion!

Link to comment
Share on other sites

Hello all!

 

My name is Kristen and I'm the supervisor of the hematology/coagulation department at our laboratory.  I have about 9 years experience as a generalist on several shifts, but I am brand-spankin' new to this role.

 

We have had issues with our sed rate analyzer for a few weeks.  The company is sending us a new one tomorrow (same model, etc...). 

 

I realize things like accuracy, precision, and analytical range must be established.  However, is it any different for this analyzer since it is the same method, model, etc..?  And, how would your facility go about establishing these things in a situation like this?

 

I was planning on running QC on both the old and new analyzer for at least one week - this should prove accuracy sufficiently.  I was also going to run a batch of patient samples as a comparison on the old vs the new analyzer...

 

How do you prove analytical range?  Would you try to establish reproducibility as well?

 

I'd appreciate any help on this.  My resources are, unfortunately, vague.

"We have had issues with our sed rate analyzer for a few weeks"   Issues? If the issues are so bad that you need a new one why use the old one for comparison?

Link to comment
Share on other sites

"We have had issues with our sed rate analyzer for a few weeks"   Issues? If the issues are so bad that you need a new one why use the old one for comparison?

 

Grippy,

 

The only "issues" that were occurring with our analyzer were happening with the barcode reader and the ability to manually type in patient IDs.  The analyzer itself has been, and still is, functioning just fine (ie, QC has been within limits, etc..).

 

Obviously if the problem with the original instrument was analytical, I would not be using it for anything at this point - including comparison studies ;)

Link to comment
Share on other sites

"We have had issues with our sed rate analyzer for a few weeks"   Issues? If the issues are so bad that you need a new one why use the old one for comparison?

 

Grippy,

 

The only "issues" that were occurring with our analyzer were happening with the barcode reader and the ability to manually type in patient IDs.  The analyzer itself has been, and still is, functioning just fine (ie, QC has been within limits, etc..).

 

Obviously if the problem with the original instrument was analytical, I would not be using it for anything at this point - including comparison studies ;)

Now I know.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.