Does anyone have a cooler validation procedure for small coolers used with wet ice and a thermometer to send blood to surgery? Any help woul dbe appreicated. Thanks in advance.

Gretchen

GRETCHEN.S.HICKMAN@osfhealthcare.org

  • Replies 9
  • Views 5.7k
  • Created
  • Last Reply

Top Posters In This Topic

comment_20538

I don't have such a process, but also would be interested if someone does - seems AABB new std 5.1.8.2.1 is asking for this?

comment_20558
I don't have such a process, but also would be interested if someone does - seems AABB new std 5.1.8.2.1 is asking for this?

Margaret: If you will post your email address or send me a Private Message here on BBTalk, I will email my cooler validation procedure to you.

Gretchen: I will email my procedure to you this afternoon.

Donna

comment_20701

Validation of small coolers has to meet the following criteria:

Whatever media is used to cool the contents, we have to prove that it will stay

within certain limits during transport.

Wet Ice in this case is ± 0 ºC as long as the Ice is melting. Just above the bloods freezing point.

Once molten, the temperature will rise.

To validate the cooler, we also need to know, the maximum time it can be in transport.

Better insulation means longer temp. stability.

Another variable is the exposure to ambient temperature.

Now, to validate the cooler.

Insert a documented amount of wet ice in the cooler box.

More Ice, means longer stability.

Insert a calibrated "logger" where the samples are going to be stored.

Place it in a room that mimics the maximum exposure temperature during transport.

Leave it exposed for the maximum time the cooler could be in transport.

Next, take the readings from the logger, to prove, the cooler contents stayed within

the limits.

Document these findings with the coolers operations log, and depending on visual damage

repeat this whenever necessary.

Usually, if no damage to the box is visible, once a year.

Damaged insulation can shorten the temperature stability of the box.

This can be easily checked, by pressing the inner lining of the box, and to see if there is any play

between the lining and the internal foam.

Now, the SOP for transport could be as followed. This could be used for any blood transport.

Blood samples removed from equipment "X" at " Date and Time". Actual temperature can be retrieved

from monitoring system

Blood samples delivered at "Y" at " Date and Time" in box "Z". ( Signature )

Data from logger "A" that was with the transport will be added to transport log file, to prove temp. stability.

I have been testing the Elpro Libero Logger, and this little device could be perfect for the job.

After delivery, it will give you a pdf with the readings or a graph, and this can be filed with the

transport documentation.

Hope, this can be of help.

comment_20702

One thing I forgot to mention, is the fact, that using wet ice for

transport is not optimal, unless the ice is seperated from the blood

so it cannot touch the samples.

comment_20717

Please don't advertise in the forums.

Thanks for your understanding.

Edited by Cliff
Advertising removed by admin

comment_20718

Hello Hodag,

Buy...buy...buy... is easy.

They already have coolers, so let people validate what they have, if possible.

And..... even Thermosafes have to be re-validated after 1/2 or 1 year, or in case damaged.

It would be nice, if certain people on this forum would not just try to promote their products.

No offence.

  • 3 months later...
comment_23948

Donna,

Could I also have a copy of your cooler validation procedure? I would really appreciate it. Thanks!

Lisa

lisa.hooke@alvaradohospital.com

comment_23952

Could I please have a copy of the cooler validation to compare with my procedure?

Thank you

khosey@wvhcs.org