If I understand what you are asking I think that is basically how I do our "High/Low" temp checks.....put the container of glycerin (with the probe in it) in ice cold water until the alarm sounds then warm it all up again until it goes off on the high end. This reproduces what would happen for a unit - right!?!

Link to comment
Share on other sites

Whoops! Looks like I probably misunderstood scswain when I first read the post. (I guess the term "heat" instead of "warm-up" got me on the wrong track.)

Of course Janet is correct. A recommended procedure for testing the blood storage alarms can be found in the AABB "Technical Manual", 15th edition, starting on page 823.

Link to comment
Share on other sites

This is commonly called a "fire and ice" test and there are a few ways of doing it but the basic description is correct. You warm the probe until the unit alarms and cool it till it alarms. If you are just doing a static temperature check (not an alarm system) you can calibrate it at a single point in the range for which that probe is used. So, for example, you can calibrate a probe in a 1 - 6 degree refrigerator at 4 degrees and be just fine.

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.