Jump to content

sullivanlawyer

Members
  • Joined

  • Last visited

  • Country

    United States
  1. sullivanlawyer started following Neil Blumberg
  2. Hi David, you usually respond quickly. I hope you are a Covid safe. michele
  3. Hi David, I am struggling with a method to method comparison for apheresis platelet counts. Please see below. I wrote this for my data eval and determination of acceptance. ● An acceptable bias or allowable total error of ±7.0% for platelet count was set for this method to method comparison. The mean of the Abbott Cell-Dyn method is 1.5408 with a standard deviation of 0.2248. It is 95% confident that the sample mean falls between 1.47199 and 1.60961 or ±4.5%. The mean of the Coulter DxH 800/600 is 1.5397 with a standard deviation of 0.2376. It is 95% confident that the sample mean falls between 1.467 and 1.612 or ±4.7%. The difference between the two methods are within allowable error for 41 of 41 samples tested. I did a method to method on EP evaluator. Does this sound correct to you?
  4. Hello, Does anyone have a method to method SOP/form they would like to share for transfusion services? Michele

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.