Hi!  Wondering if someone can help me.  My BioMed Team only calculates g force on our centrifuges.  When we had the old Ortho Centrifuge, the User Manual Stated on the second page " The OLD MTS Centrifuge™ Model 5150-60 will reach a speed of 895 ±25 RPM. (revolutions per minute); 80-90 RCF (Relative Centrifugal Force)"  However, I am unable to find this information anywhere for the 2-in-1 Centrifuge Incubator Model ( I am attaching image).  Now, I have a calculation to convert, and I did measure the radius.  But here's the thing:  I'm not 100% sure if I am measuring the radius correctly,  and 2- I would love to have something in writing from Ortho, like in the previous manual.  Does anyone have any information, or can direct me somewhere?  I reached out to Ortho, and not only do they not know the g force in rcf, they also don't know their centrifuge radius either.

Thank you to anyone that can hep!! 

 

g = (1.118 x 10-5) x R x S2

Where g is the relative centrifugal force , R is the radius of the rotor in centimeters and S is the speed of the centrifuge in revolutions per minute (RPM). The follow table is a simple reference for common rotor sizes and their RCF values in units of times gravity (x g).  For example, an experiment requires a sample to centrifuged at 5,000g and the centrifuge that is available has a rotor of 9cm, using the table below, the centrifuge needs to be run at a speed of 7,000rpm as this gives an RCF of 4930 x g.  Using the formula the precise speed would be 7,049rpm.

The OLD MTS Centrifuge™ Model 5150-60 will reach a speed of 895 ±25 RPM. (revolutions per minute); 80-90 RCF (Relative Centrifugal Force) 

The New Ortho-Clinical Diagnostic’s 2 in 1 Centrifuge-Incubator Workstation is 1032 rpm + 10 rpm which is ??

ortho.jpg

centrifuge radius.jpg

Link to comment
Share on other sites

  • 4 weeks later...

I find it odd that your biomed team only can give a  calculated g (RCF). How would they go about doing that? Most biomed departments can only measure rpm using tachs. 

Link to comment
Share on other sites

Create an account or sign in to comment

You need to be a member in order to leave a comment

Create an account

Sign up for a new account in our community. It's easy!

Register a new account

Sign in

Already have an account? Sign in here.

Sign In Now
  • Recently Browsing   0 members

    • No registered users viewing this page.
  • Advertisement

×
×
  • Create New...

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.