  1. When you write your PQ do you include testing the maximum number of units that can be stored in the refrigerator or freezer? Currently we monitor empty and then with at least 20 units per shelf. The FDA investigator indicated that we need to validate for the maximum number of products we could store in any one unit. I'm not sure how to tackle this so any advice would be appreciated!

  2. Our facility uses several of the procedures people have mentioned previously we use secure transactions and have the units returned if the canister is not retrieved within 5 minutes and validate the trip time and temperatures.

    In addition to these we also include LDH and K+ levels in our validation for PRBC products and include a small bag of wet ice with every refrigerated product sent via pneumatic tube.

  3. How is everyone accessing the acceptability of washed/deglyc units if they are returned to the Blood Bank? We routinely place a Hemotemp on every unit that is issued but because washed/deglyced units are usually not be cold enough (4° C for 10 hours per the package insert) the Hemotemp will not activate correctly. I would really appreciate any suggestions.

×
×