Hello,

We are currently using diluted albumin as our control for Rh testing. I want to introduce Gamma Clone control made by Immucor since our Anti-D reagent is Gamma Clone. Do I need to validate or do any kind of comparison studies before I go live with the reagent. I am sure manufacturer completed all the studies but, I was wondering if we need to do any testing before I replace our Albumin control with Gamma clone control. Thanks in advance for taking time to answer my question.

  • Replies 2
  • Views 1.2k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • I just recently switched from 6% albumin to Immucor's monoclonal control.  I validated it by running a certain number of Rh Positive and Rh Negative patients performing Weak D testing as well.  We als

comment_71765

I just recently switched from 6% albumin to Immucor's monoclonal control.  I validated it by running a certain number of Rh Positive and Rh Negative patients performing Weak D testing as well.  We also tested it with our Daily QC because we used the 6% albumin as our negative control.  I really debated about the necessity in validating a Control product but decided just to perform a short validation just in case anyone ever asked.:)

Edited by mld123

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.