I am trying to create the conditions by which we can perform our Cord Blood Testing via automation.  As we all know, cord blood specimens are not great.  We were using a sterile screw top tube that they had used years ago, but in my efforts to see if we could automate it, we were able to locate sterile pink EDTA tubes for them to collect in.  However, still getting them with clots.  They found that if they filled them only 1/2 full (so quicker to get cap back on), the clotting was not as bad or frequent....but still no consistency.  Our automation will run testing on Cord Specimens (Erytra), but we will not start it until when and if we can get them non-clotted (or at least where we can use an applicator stick and maybe just be taking out 1 small clot).

Anyone out there successfully performing cord blood testing via automation?  I can't imagine any analyzer would accept a clotted specimen so if you are, please share with me what your success is in getting non-clotted cord specimens. :)

Thanks,

Brenda Hutson, MT (ASCP)SBB

  • Replies 14
  • Views 2.8k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • We do. We also have Erytra and wadiana.  Our cord blood samples are collected in either purple or pink edta tubes. All specimens get spun when received. If a cord sample needs to have testing per

  • Laurie Underwood
    Laurie Underwood

    We run cord blood samples collected in a purple top EDTA on the TANGO. Check for and remove any clots with applicator sticks and centrifuge for 5 minutes before testing. Has been working fine.

comment_69199

We do. We also have Erytra and wadiana. 

Our cord blood samples are collected in either purple or pink edta tubes. All specimens get spun when received. If a cord sample needs to have testing performed, any clots are removed with applicator sticks and respun prior to testing. 

comment_69200

Our cord blood samples are collected in red tops. They arrive in BBK usually with one giant clot.  We remove the clot, centrifuge and test on ProVue.  We stopped using EDTA samples because they usually had numerous small clots that were very time consuming to remove.

Ah, so give up trying to "receive" non-clotted specimens.....just have to create them on our own.  Either of you have issues because not all clots were removed sufficiently?

Thanks,

Brenda

comment_69223

We test our cord bloods on the Echo by removing an aliquot of the cells from the EDTA tube.  We can usually get clot-free cells this way.  Any small clots are removed prior to analyzing.

comment_69228
1 hour ago, Brenda Hutson said:

Ah, so give up trying to "receive" non-clotted specimens.....just have to create them on our own.  Either of you have issues because not all clots were removed sufficiently?

Thanks,

Brenda

Rarely

comment_69239

We are running cord blood samples on an Echo for type, DAT and weak D. At first we were running cells collected from a red top tube - usually plenty of fall out cells in the bottom, but we had to wait for them to clot. I finally convinced delivery to collect in EDTA tubes using the sterile tube kit mentioned above. We very carefully remove any and all clots with applicator sticks and centrifuge. Working very well.

comment_69244

We run cord blood samples collected in a purple top EDTA on the TANGO. Check for and remove any clots with applicator sticks and centrifuge for 5 minutes before testing. Has been working fine.

comment_69251

We remove an aliquot from the primary container and rim the sample for clots prior to centrifugation. We use the Ortho Vision analyzers. Prior to that we used the ProVue.

comment_69265

We also put our cord bloods on the Tango after removing clots.  Works great.

comment_69286

We get our cords collected in a sterile Screw top tube, no anticoagulant.  We've been testing on our analyzer for years now, 1st ProVue and now Vision.  For some reason, after these clot and retract there is very abundant serum with lots of free RBC's.  We just pull a small aliquot of serum and RBC's off, ream for clots (which are rare using this method) and then centrifuge.

  • 1 month later...
On ‎3‎/‎23‎/‎2017 at 1:27 PM, Sandy L said:

We get our cords collected in a sterile Screw top tube, no anticoagulant.  We've been testing on our analyzer for years now, 1st ProVue and now Vision.  For some reason, after these clot and retract there is very abundant serum with lots of free RBC's.  We just pull a small aliquot of serum and RBC's off, ream for clots (which are rare using this method) and then centrifuge.

So are you placing that small amount of serum and RBCs, in a separate tube for testing?

Brenda

comment_69844
On ‎3‎/‎21‎/‎2017 at 4:51 PM, Laurie Underwood said:

We run cord blood samples collected in a purple top EDTA on the TANGO. Check for and remove any clots with applicator sticks and centrifuge for 5 minutes before testing. Has been working fine.

Do you have a protocol for Cord testing on the Tango. How did you validate the DAT coombs?

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.