Hi everyone,

We have seen some patients that are demonstrating anti-K with PEG only.  We repeat with Gel and tube LISS and results are negative.  We change our incubation times and still negative, we have used K homologous cells in gel and still negative.  Anyone have any hints as to why this is happening.

Chris

  • Replies 5
  • Views 2.3k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • Could be either. In my experience, PeG is a bit more sensitive than gel. When we had nebulous reactions in gel, we resorted to PeG to ID the antibody. I think it's always valuable to remember tha

  • PEG-only antibodies are well known, but it's common that they can also be detected in one of the other very sensitive methods (Gel or Solid Phase). The LISS method is less sensitive than PEG and it's

  • If you read the manufacturer's instructions for Ortho Gel it says: "In some low ionic strength test systems, certain antibodies, such as anti-E and anti-K, have been reported to be nonreactive."

comment_67187

PEG-only antibodies are well known, but it's common that they can also be detected in one of the other very sensitive methods (Gel or Solid Phase). The LISS method is less sensitive than PEG and it's not surprising that you can't detect it there.

It is possible that the PEG reagent contains a chemical that is critical to the reactivity of your patient's particular antibody. There are lots of documented examples of "chemical-X-dependent" antibodies. The absence of the specific chemical from the LISS test and the Gel system would render it nonreactive/undetectable.

Trying another manufacturer's PEG reagent (presumably with a different formulation) might clarify the situation.

comment_67190

If you read the manufacturer's instructions for Ortho Gel it says: "In some low ionic strength test systems, certain antibodies, such as anti-E and anti-K, have been reported to be nonreactive."

comment_67192
1 hour ago, goodchild said:

If you read the manufacturer's instructions for Ortho Gel it says: "In some low ionic strength test systems, certain antibodies, such as anti-E and anti-K, have been reported to be nonreactive."

Good point. I know that the Gamma/Immucor PEG reagent available in the US does have low ionic properties. Don't know about other suppliers, but it's common for reagents like these to be very similar.

  • Author
comment_67196

But we are seeing anti-K and anti-E in gel; so is it the genetics of these antibodies or is it that they are so weak that gel just does not pick them up? 

 

comment_67200

Could be either. In my experience, PeG is a bit more sensitive than gel. When we had nebulous reactions in gel, we resorted to PeG to ID the antibody.

I think it's always valuable to remember that there is no one system that will detect all antibodies, even within the same specificity - that's just life (or maybe I should say, that's just biology).

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.