We have a Vision coming, and I was wondering about getting up and running and validated. I am looking to avoid problems and delays in putting into use. Has anyone interface it with Cerner yet?

  • Replies 39
  • Views 16.1k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • LKSchroed
    LKSchroed

    Our Vision is to be installed tomorrow (May 24).  One tech goes to training on June 13.  The inhouse training starts June 20.   We have Cerner Milennium, but will also be using middleware (DataInnovat

  • butlermom
    butlermom

    LKSchroed, we are getting ready to order 2 Visions and we also have Cerner.  I'm guessing it will be mid to late summer by the time we are up and running.  Let's keep in touch as we both go through th

  • LKSchroed
    LKSchroed

    Everything with the Vision installation is on schedule.  We have a conference call scheduled with Ortho personnel to talk more about in-house training, validation schedule, and interfacing with C

comment_62312

ORTHO will probably give some insight and direction for you.  Has your lab selected a key person to send for training?

  • Author
comment_62319

Yes, I have two techs to send for training.  Originally I was told training would be in-house.  Now I hear that they will travel somewhere.

comment_62337

It would be great if you could periodically post your progress and experiences as you go through this process, LKSchroed.  I am sure that many of us would be interested in how things with the Vision go for you.  (Thanks, in advance!)

 

Donna

comment_62343

I went to training last week and we are doing the setup right now. I'm very impressed. We are Meditech CS 5.6 and have had a Provue for about 9 years.

 

Gerald

  • 5 months later...
  • Author
comment_65267

Sad to say we still don't have our Ortho Vision.  Not an Ortho problem, just a delay by my hospital in spending the money.  Would love to hear from anyone who has a Vision up and running.

thanks!

  • 1 month later...
comment_65667

LKSchroed, we are getting ready to order 2 Visions and we also have Cerner.  I'm guessing it will be mid to late summer by the time we are up and running.  Let's keep in touch as we both go through this process. 

comment_65674

We currently have Immucor Neo and are getting proposals form both Immucor and Ortho for our contract renewal. We also are Cerner Millennium so I would be interested in how the Ortho interface process goes for the Vision.

  • 2 weeks later...
  • Author
comment_65847

Our Vision is to be installed tomorrow (May 24).  One tech goes to training on June 13.  The inhouse training starts June 20.   We have Cerner Milennium, but will also be using middleware (DataInnovations.) 

 

comment_65849

I'm curious about your planned use of middleware. The rest of our laboratory uses middleware (and specifically DataInnovations).

  • Author
comment_65943

Everything with the Vision installation is on schedule. 

We have a conference call scheduled with Ortho personnel to talk more about in-house training, validation schedule, and interfacing with Cerner.

But the Vision looks great sitting in the BB lab!

Will post progress and if any issues arise.

comment_65944
10 hours ago, LKSchroed said:

Everything with the Vision installation is on schedule. 

We have a conference call scheduled with Ortho personnel to talk more about in-house training, validation schedule, and interfacing with Cerner.

But the Vision looks great sitting in the BB lab!

Will post progress and if any issues arise.

Thank you for keeping us updated.

  • 4 weeks later...
  • Author
comment_66275

Ortho Vision has been installed.  We are training techs on use.  I am writing the validation process as fast as possible along with required procedures. 

We have hit a snag with connecting Ortho Vision to DataInnovations to Cerner Millenium.  Right now Cerner doesn't like the info coming from the Vision via DI.  Not sure yet if its just a matter of mapping or what the problem is.

So far the techs being trained think it is great.

comment_66326

Do you find the noise level when the carousel turns any problem?  I am wondering if we will need to consider that in where we locate the Vision we will be getting this fall.

  • 2 weeks later...
comment_66539
On Monday, May 23, 2016 at 2:12 PM, LKSchroed said:

Our Vision is to be installed tomorrow (May 24).  One tech goes to training on June 13.  The inhouse training starts June 20.   We have Cerner Milennium, but will also be using middleware (DataInnovations.)

 

Is there a reason you chose to use a middleware with the Vision?

Thanks.

comment_66544

I'm also curious. We didn't set up middleware with our ProVue to Meditech interface.

  • 3 months later...
comment_67582

This does not answer the original question, since we are interfaced with Softbank, but I am going to warn you to the potential problems that are in store for you:

We have had two of these instruments up and running since August.  Here are the problems we have:

1. QC is a pain.  Half the time it does not pass because the instrument can't interpret the image. You can't change the result, so you have to rerun the QC. Sometimes you have to rerun it multiple times. It can take half the day just to get the QC to work and there have been times that the QC has timed out on both instruments and we have had to resort to manual testing. Not good for a busy hospital like ours. There is a software update in the works that will allow techs to manually interpret the image, but I feel this is a bandaid on a bigger problem, which is the sensitivity of the instrument to artifacts (see below).

2. The instrument is extremely sensitive, so even dust will cause it not to be able to interpret images. Techs manually have to change the grade. 

3. After centrifugation of the cards, we will sometimes see streaks of cells up both sides of the well, but not in the center. The instrument will interpret this as positive, but we have learned through experience that if we repeat with manual gel testing, it will be a nice negative result.  We never saw this happen with the ProVue and I wonder if it is a problem with the centrifuge. Perhaps if it used a balance card, we wouldn't have this problem. It's just my theory.

4. You can't put partially used cards back on the instrument.  It causes a lot of waste.  Also having to repeat QC so many times is wasteful.

5. Aside from the above problems, we love the instrument. 

comment_67596
On ‎11‎/‎12‎/‎2016 at 7:13 AM, Kathy said:

This does not answer the original question, since we are interfaced with Softbank, but I am going to warn you to the potential problems that are in store for you:

We have had two of these instruments up and running since August.  Here are the problems we have:

1. QC is a pain.  Half the time it does not pass because the instrument can't interpret the image. You can't change the result, so you have to rerun the QC. Sometimes you have to rerun it multiple times. It can take half the day just to get the QC to work and there have been times that the QC has timed out on both instruments and we have had to resort to manual testing. Not good for a busy hospital like ours. There is a software update in the works that will allow techs to manually interpret the image, but I feel this is a bandaid on a bigger problem, which is the sensitivity of the instrument to artifacts (see below).

2. The instrument is extremely sensitive, so even dust will cause it not to be able to interpret images. Techs manually have to change the grade. 

3. After centrifugation of the cards, we will sometimes see streaks of cells up both sides of the well, but not in the center. The instrument will interpret this as positive, but we have learned through experience that if we repeat with manual gel testing, it will be a nice negative result.  We never saw this happen with the ProVue and I wonder if it is a problem with the centrifuge. Perhaps if it used a balance card, we wouldn't have this problem. It's just my theory.

4. You can't put partially used cards back on the instrument.  It causes a lot of waste.  Also having to repeat QC so many times is wasteful.

5. Aside from the above problems, we love the instrument. 

After that laundry list what's left to love?  I do like the user interface. . .

comment_67605
On ‎11‎/‎12‎/‎2016 at 6:13 AM, Kathy said:

This does not answer the original question, since we are interfaced with Softbank, but I am going to warn you to the potential problems that are in store for you:

We have had two of these instruments up and running since August.  Here are the problems we have:

1. QC is a pain.  Half the time it does not pass because the instrument can't interpret the image. You can't change the result, so you have to rerun the QC. Sometimes you have to rerun it multiple times. It can take half the day just to get the QC to work and there have been times that the QC has timed out on both instruments and we have had to resort to manual testing. Not good for a busy hospital like ours. There is a software update in the works that will allow techs to manually interpret the image, but I feel this is a bandaid on a bigger problem, which is the sensitivity of the instrument to artifacts (see below).

2. The instrument is extremely sensitive, so even dust will cause it not to be able to interpret images. Techs manually have to change the grade. 

3. After centrifugation of the cards, we will sometimes see streaks of cells up both sides of the well, but not in the center. The instrument will interpret this as positive, but we have learned through experience that if we repeat with manual gel testing, it will be a nice negative result.  We never saw this happen with the ProVue and I wonder if it is a problem with the centrifuge. Perhaps if it used a balance card, we wouldn't have this problem. It's just my theory.

4. You can't put partially used cards back on the instrument.  It causes a lot of waste.  Also having to repeat QC so many times is wasteful.

5. Aside from the above problems, we love the instrument. 

ABOUT #1 - just wondering why can't you change a QC result?  Are you allowed to change patient reaction results?

comment_67611
On 6/29/2016 at 2:06 PM, LKSchroed said:

Ortho Vision has been installed.  We are training techs on use.  I am writing the validation process as fast as possible along with required procedures. 

We have hit a snag with connecting Ortho Vision to DataInnovations to Cerner Millenium.  Right now Cerner doesn't like the info coming from the Vision via DI.  Not sure yet if its just a matter of mapping or what the problem is.

So far the techs being trained think it is great.

We have had a terrible time with the DI product, Instrument Manager.  The Lantronix box has failed twice (the network guy didn't plug it into a surge protector), and the mapping is a pain, even with direct help from DI.  

comment_67673

when I interfaced ProVue to cerner classic...it was a pain and took long time. make sure the names matches as it is require din Vision, DI and cerner Milennium.

middleware requirement is based on LIS system. Cerner needs DI. I thought soft doesn't need DI to connect to ProVue it may be different with Vision.

comment_67682
On ‎11‎/‎14‎/‎2016 at 9:49 AM, R1R2 said:

ABOUT #1 - just wondering why can't you change a QC result?  Are you allowed to change patient reaction results?

You can't change a QC result because that's how the Ortho people programmed the instrument. You are allowed to change patient reaction results.  The software update will allow users to change the QC result.

comment_67696
20 hours ago, Kathy said:

You can't change a QC result because that's how the Ortho people programmed the instrument. You are allowed to change patient reaction results.  The software update will allow users to change the QC result.

Interesting change. Immucor also doesn't allow changes to QC, which is a pain when you can plainly see that the problem is an artifact in a well.

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.