I would like to know the practical methods with details for QC of anti-human globulin, anti- A, Anti-B and  Anti-D reagents after buying a new trade mark or change in lot numbers. 

  • Replies 5
  • Views 2.1k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • David Saikin
    David Saikin

    If you use FDA licensed reagents (in the USA) - all I would do is verify that they pass my routine QC of such reagents.  There is no need to do parallel testing (at least in my opinion).

  • David Saikin
    David Saikin

    If you are having problems (? lack of reactivity, false positive) I would look at how you are storing the reagents, are they out of the refrigerated environment for long periods of time, are your tech

comment_61539

If you use FDA licensed reagents (in the USA) - all I would do is verify that they pass my routine QC of such reagents.  There is no need to do parallel testing (at least in my opinion).

  • Author
comment_61553

If you use FDA licensed reagents (in the USA) - all I would do is verify that they pass my routine QC of such reagents.  There is no need to do parallel testing (at least in my opinion).

we do not use FDA licensed reagents and some time we have encountered with some problems using the reagents. 

comment_61555

we do not use FDA licensed reagents and some time we have encountered with some problems using the reagents. 

If you are having problems (? lack of reactivity, false positive) I would look at how you are storing the reagents, are they out of the refrigerated environment for long periods of time, are your techs contaminating them somehow, are your serofuges set at the correct time (and have you done the testing to validate the correct spin times for each reagent), is your saline/diluent contaminated?  As you can see there are many variables to consider - HOWEVER, the most important and usually the culprit is the humans.

 

You would normally qc your anti-A, -B, -D reagents with a positive and negative cell.  For your anti-IgG I would use sensitized and non-sensitized rbcs..

 

Don't buy from the low bid if these are the reagents giving you problems.  I always try to purchase the best reagents I can find for blood bank testing.

Edited by David Saikin

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.