comment_59291

We've seen the same issues since December 2014, and found thorough cleaning of the probe block helpful, sometimes cleaning the washer manifold &/or changing the indicator cells &/or changing the screening cells helpful also.

Immucor has become more helpful and efficient at shipping replacement reagents quickly. 

I find the Echo Report  for  Strip Performance or Reagent Performance helpful. 

As the percentage of equivocal reactions rises to 2% you need to intervene or take action. 

Once it exceeds 2%, and you've done the maintenance the reagents are readily replaced.

  • Replies 25
  • Views 8.1k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • But remember when you compare two totally different methods you are going to have discrepancies (Not all methods will detect all antibodies all of the time). While we would like to have 100% correlati

  • We have had these problems intermittently with the ECHO.  This will be our third winter to have it and these non-specific reactions seem to increase in numbers around the middle of February.  I live i

  • We had been seeing this problem for several months, different lots and having to use our backup gel method and report those results.  The last lots of screening cells we didn't call as we switching to

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.