Hi all,

 

I'm trying to re-vamp our preventative maintenance/equipment function check process.  I can't seem to find the following in any of the CAP checklists (lab general, all common, or hematology/coags):

 

  • How often must centrifuge timers be checked?
  • How often must centrifuge rpms be checked?
  • How often is it required to have microscopes PM'ed?

Wondering with what frequency the rest of you perform these.  Currently, we are checking timers and rpms quarterly and microscopes are serviced semi-annually.  Seems like it may be overkill...

  • Replies 8
  • Views 5.3k
  • Created
  • Last Reply

Top Posters In This Topic

  • Author
comment_53906

In addition - what about checking the temps of waterbaths against certified thermometers?  Can't seem to find anything on that either... Also currently doing this quarterly.

comment_53907

We do centrifuge RPMs and timers quarterly. Thermometer checks (including water bath) annually. Microscopes every 6 months.

We're not CAP, so not sure if these agree with CAP. We are accredited by the Joint Commission and NY State Dept of Health.

  • Author
comment_53912

Thanks, tbostock!  Always interesting to see what other people are doing.  Hoping I get some reponses from other CAP accredited labs so we can compare =)  I see you are, in fact, doing rpm and timers quarterly.  Is this a requirement by JCAHO/NY State Dept of Health - or something you just do?

comment_53977

We are doing same for all three.

We were doing annual pm on microscope but something changed,

Was it hematology checklist? We often get our microscope

From hematogy.

comment_53994

 

Thanks, tbostock!  Always interesting to see what other people are doing.  Hoping I get some reponses from other CAP accredited labs so we can compare =)  I see you are, in fact, doing rpm and timers quarterly.  Is this a requirement by JCAHO/NY State Dept of Health - or something you just do?

 

Yes, quarterly centrifuge checks are required. The JC and NYS fall in line with AABB recommendations, which are quarterly. Technical Manual 17th edition, p. 37

  • 3 weeks later...
comment_54160

RPMS are supposed to be done on centrifuges twice a year, we use every 6 months. Timers are done quarterly, and thermometers are done semi-annually.

  • 2 weeks later...
  • Author
comment_54255

Thanks, MAGNUM.  Can you tell me where I can find this specific information in the CAP guidelines?  I've checked out the general, all common, and hematology/coag checklists - cannot find it.  Please help!  lol

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.