Any labs out there allow their techs to perform manual differential counts by high dry (50x) objective as opposed to oil immersion (100x) as a routine? An internet search would seem to indicate that some labs, by policy, do allow the use of 50x. Anyone willing to share their thoughts on this subject?

  • Replies 6
  • Views 3.8k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • Pavel Dmitrievich
    Pavel Dmitrievich

    I've always believed that oil at 100x gives the best view of detail. We look at the expected and anticipate the unexpected. Since we live in an area here  with a higher incidence of Lymes and Anaplasm

  • 50. If our lab was to use 100 we would all be spending twice the time doing diffs and much would be missed. Ever take a slide to the pathologist? They use a lower power than 50 to get the best overall

comment_52608

We have both 50 and 100x oil objectives on our scopes.   Techs use whatever they are comfortable with. No one that I know of would use a hi-dry here for a differential.

 

I usually only use 100x when looking at alot of cells, like in a bone marrow or cytospin of a chest fluid.

 

Scott

comment_52623

Same as Scott, 50 and 100x oil.  Techs count with whatever is most comfortable.

Deny

comment_52651

I've always believed that oil at 100x gives the best view of detail. We look at the expected and anticipate the unexpected. Since we live in an area here  with a higher incidence of Lymes and Anaplasma, we are always keyed to inclusions, which might not be seen at 50X. The rush to put out results, and I'm not being snarky, shouldn't preclude our giving it our best.

comment_52695

We us 50x OIL (the objective is an oil one), and go to 100x for very cellular blood smears, Bone Marrows and cytospins.

  • 4 weeks later...
comment_53031

We use oil x50. Can't see the need to do a x100 unless you are specifically looking at the chromatin of the cells. For a basic diff x50 is plenty.

 

Also if you increase the magnification to x100 it makes it more difficult to see that you have drifted from the 'right' part of the film and could actually result in a false differential. I am forever telling trainees 'too thick' or 'too thin' when I am teaching them how to do an accurate diff.

comment_53039

50. If our lab was to use 100 we would all be spending twice the time doing diffs and much would be missed. Ever take a slide to the pathologist? They use a lower power than 50 to get the best overall view. Only go to 100 if there is something that needs inspection. If someone needs 100 to do a diff send them back to the student desk

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.