comment_49427

Our cord blood screen procedure, which was in place when I took over as senior tech in blood bank, has always stated to make a 3-5% washed cell suspension to do the ABO/Rh type and then wash cell suspension an additional 8 times before adding reagents for direct coombs and a control. I think this is over kill, since washing it 4 times, then adding reagents results in a negative control. (If the control were positive, then additional washes would be necessary).

Our cord bloods are collected by nursing with a needle and syringe and placed into an EDTA purple top tube. The technical manual also states that collecting with a needle and syringe avoids contamination with Wharton's jelly and the need for additional washings. I know the reason behind the "8 washes" was to wash away the Wharton's jelly, but with EDTA tubes, I've not seen any interfering substances. Plus, I cannot find a reference to wash it 8 times, just the tech manual now stating that additional washes are not needed if collected with needle and syringe.

So, with all that being said, what does everyone else do? I'd love to change it make a 3-5% cell suspension (no washing needed to do the ABO/Rh type) and then wash the cell suspension 4 times and then add the reagents for the direct coombs and the control.

Thanks

Natalie


  • Replies 23
  • Views 10.8k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • We wash a suspension of cord red blood cells maunually 3x. From these washed cells we do an ABO and Rh type in tube. Also, from the washed cell suspension, 1 and 2 drops of it are placed into 2 respec

comment_49428

Make a 3-5% cell suspension (no washing needed to do the ABO/Rh type) and then wash the cell suspension 4 times and then add the reagents for the direct coombs and the control. There is a note that if the control is positive to repeat with 8-12 washes.

comment_49429

We do all our testing using automation unless the specimen is clotted. There is no washing involved there. If we do manual testing, we just make a cell suspension with no washing.

  • Author
comment_49432
Make a 3-5% cell suspension (no washing needed to do the ABO/Rh type) and then wash the cell suspension 4 times and then add the reagents for the direct coombs and the control. There is a note that if the control is positive to repeat with 8-12 washes.

Any reference besides current technical manual?

- - - Updated - - -

We do all our testing using automation unless the specimen is clotted. There is no washing involved there. If we do manual testing, we just make a cell suspension with no washing.

We are still manual tube testing, no automation here yet

comment_49435

If you are using monoclonal ABO antibodies, there is no need to wash.

If you are using human ABO reagents, there is a need to wash.

For the DAT, wash four times.

comment_49436

Just AABB and the current practices in the area.

  • Author
comment_49438
If you are using monoclonal ABO antibodies, there is no need to wash.

If you are using human ABO reagents, there is a need to wash.

For the DAT, wash four times.

They are monoclonal for ABO antibodies

comment_49439

Then no need to wash.

comment_49440

We only wash if we have dispcrepancies - use gel.

Even in the "old days" we only washed 4 times unless there was a problem

comment_49457

Current (old) policy had us perform the aborh and set up DAT and DAT ctl at the same time - in case DAT is needed. Make a 3-5% solution and wash all 6 tubes at the same time...seemed to save time by doing it all together....guess we don't need to...

comment_49471

We make a cell suspension and place 2 drops in 5 tubes and put them all in the cell washer. Wash 4X. We do a DAT on all newborns. Seems like a lot of effort to divide up 5 tubes regardless of the reagents. A couple of times a year out of 1500 babies do we have to wash more than 4.

comment_49533

I'm not sure if I am following the DAT descriptions correctly, but I remember having debates in the past over whether you could wash a cell suspension X number of times, place a drop of these cells in a tube and just add the antiglobulin reagent or whether you had to wash that drop of cells at least one more time to a dry button before adding the antiglobulin reagent. It seems that I recall that the manufacturer's instructions implied that you had to wash it to a dry button. What is the consensus and has anyone tested it both ways?

comment_49538

We don't centrifuge to a completely dry button. There is always a little saline left on the cells.

comment_49541

We make a 3% suspension of unwashed cells for the ABO and RH and wash a drop of cells in the cell washer 4 times for the DAT. If our ABO and RH looks grainy then we might wash the cells 4 times and repeat.

comment_49544

Malcolm, are you talking about a single drop of cells washed the last time still having residual saline (what I was loosely calling a dry button--even though you had not blotted off the residual saline or anything) or are you talking about a drop of cells that is around 3% cells in saline. Or does it matter? The instructions always say, "Decant completely after last wash" which suggests to me that they don't intend for us to use a drop of 3% cell suspension to which we add AHG. But maybe that statement is in a different context and I am being overly narrow in my interpretation. Now there is an easy SBB project?!?!!

comment_49546

Sorry Mabel, my fault for not being clear. I meant just residual saline (not blotted off), rather than 3% cells in saline.

comment_49549

we make 3%(this is with measured saline and packed cells) suspension from CORD pack cells and then take one drop of that suspension and wash it X3 in cell washer, then add 2 drops if antiglobulin reagent, spin, read, if negative read microscopically.

comment_49553
Then no need to wash.

No need to wash the RBC.

I stopped washing when we went over to gel cards but my colleagues began to complain. :)

comment_49561

We routinely test unwashed cord blood samples on ProVue without any problems.

comment_49725

We do our cord testing in tubes. I wash once, just to make sure all clots etc are out (nurses are notorious for not inverting the EDTA tubes). This initial suspension is used for ABO and Rh testing. If the Rh needs to be taken to IgG, I then wash a drop of the suspension 4 times, decanting to the "dry button" each time. The same is done for the DAT.

comment_49734

We also test unwashed cord blood samples on ProVue with no problems.

We routinely test unwashed cord blood samples on ProVue without any problems.
comment_49804

We wash a suspension of cord red blood cells maunually 3x.

From these washed cells we do an ABO and Rh type in tube.

Also, from the washed cell suspension, 1 and 2 drops of it are

placed into 2 respective tubes, wash 3-4 times in the cells washer

and then add 2 drops of Anti-IgG to eaxh tube for the DAT.

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.