I am currently in the process of purchasing a new centrifuge for our EDTA specimens. This is the first time I have had to validate this type of equipment and am unsure of where to start. Any tips?

Thanks.

  • Replies 4
  • Views 2.1k
  • Created
  • Last Reply

Top Posters In This Topic

Most Popular Posts

  • Hi azizka71, What is your email, and I will send you my SOP? Dalene

comment_48215

Include the IQ, OQ, PQ:

IQ: install it the way that the manufacturer requires and document that

OQ: make sure it operates the way the manufacturer says it does. Check the RPMs, the timer, if it beeps when they say it should, etc.

PQ: make sure it performs the way YOU want it to. Spin specimens and run them in comparison to your current centrifuge to make sure you get the same results. Don't have to go crazy on this one, since you are really just making sure that the RPMs and spin time of the centrifuge give you what you need to do accurate testing.

  • 7 months later...
comment_52081

Hi, how can calibrate the centrifuge to separete the most platelets from a unit of whole blood ?

 

Thanks

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.