Do I meet the CAP requirement of QC-ing a new test kit (example: Fetalscreen) at the same time as a patient on 1st day of use? Do I have to check the old lot of QC (pos/neg) with the new anti-D reagent in the kit and vice versa?

I am thinking of when I worked in hematology and we ran a pos/neg patient with old and new kits (ex: Sed rate, FDP). But at my current facility we don't get very many Fetalscreens so I don't think I could accomplish this, as long as we are not combining kits of different lots?

Any input is appreciated...

  • Replies 6
  • Views 2.4k
  • Created
  • Last Reply

Top Posters In This Topic

comment_44129

I qc my fetal screen using the new anti-D with the old kit controls. This has never been a regulatory problem and I have used these kits for decades.

comment_44196

We only QC the new lot with the new controls that come with the kit. I don't think there is any requirement to do this crosschecking with controls from the old kit. In any case CAP TRM.31350 states "If there are multiple components of a reagent kit, the laboratory uses componenets of reagent kits only within the kit lot unless otherwise specified by the manufacturer."

Bill

comment_44199

the day of receipt we run pos & neg controls(from the kit itself) from old lot and new lot (lot to lot release). (we do not interchange component from each kit---because of the CAP).

  • Author
comment_44212
We only QC the new lot with the new controls that come with the kit. I don't think there is any requirement to do this crosschecking with controls from the old kit. In any case CAP TRM.31350 states "If there are multiple components of a reagent kit, the laboratory uses componenets of reagent kits only within the kit lot unless otherwise specified by the manufacturer."

Bill

I think I agree, Bill. From what I recall, CAP is more concerned with PATIENT results with old, new lots (no significant variation in result with either lot). I remember it being a pain with urine dipsticks doing lot to lot comparisons. If we don't get Fetalscreen tests for several days or a week, the samples are no longer "stable". Plus we hardly see positive fetal screens (maybe 2 a year). Thanks everyone for the input.

Colleen Hinrichsen, MT (ASCP) SBB

comment_44216

I have considered lately running the old controls with the new Ortho kit because we have had some times when the positive control is too weak. If we run the old controls while they are still in date and they work fine but the new pos control doesn't, then we know that it is the control failing rather than the reagents failing. In this case, we would be using the old controls as an "unknown" rather than "mixing kits."

comment_44289
We only QC the new lot with the new controls that come with the kit. I don't think there is any requirement to do this crosschecking with controls from the old kit. In any case CAP TRM.31350 states "If there are multiple components of a reagent kit, the laboratory uses componenets of reagent kits only within the kit lot unless otherwise specified by the manufacturer."

Bill

Same here.

Donna

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.