To be in compliance with the TRM's 32216 and 32232, I would like to know what other facilities do for their MTS dispensers for the Dil 2 and Dil 2 Plus containers used for the ortho gel system? How are these checked for calibration and who does it? As for accuracy, what methods are being used to check accuracy? How often do you check for accuracy of the dispensers?

  • Replies 6
  • Views 3.5k
  • Created
  • Last Reply

Top Posters In This Topic

comment_11534

We bleach ours once a week and as we are running the DI water through it for the rinse we squirt 10 squirts (5 mL) into a 10 mL graduated cylinder. Tolerance limits are +/- 0.5 mL.

comment_11537

Per the insert, we rinse ours 1mL dispenser weekly with Alcohol and DI water. We test 10 dispenses at 10.0 + 0.5 mL after each cleaning.

comment_11538

We follow manuf. insert. We clean our 0.5 ml and 1 mL dispensers with 10% alcohol and Distilled water every week.

comment_11577

Remember, minute amounts of bleach can destroy the S antigen. Hence the alcohol the manufacturer suggests, I imagine.

Create an account or sign in to comment

Recently Browsing 0

  • No registered users viewing this page.

Important Information

We have placed cookies on your device to help make this website better. You can adjust your cookie settings, otherwise we'll assume you're okay to continue.

Configure browser push notifications

Chrome (Android)
  1. Tap the lock icon next to the address bar.
  2. Tap Permissions → Notifications.
  3. Adjust your preference.
Chrome (Desktop)
  1. Click the padlock icon in the address bar.
  2. Select Site settings.
  3. Find Notifications and adjust your preference.